Re: ki-67 stain permeabilization?

From: Derek Davies <derek.davies@cancer.org.uk>
Date: Wed Jul 02 2008 - 05:18:58 EDT
Hi Alessandro,

We have had good results simply from 70% ethanol permeabilisation. I dimly
recall that we also tried paraformaldehyde (1%	then 0.1% Triton X-100 and
got roughly similar results but the ethanol was more convenient). It may
depend on whether you also want to detect surface antigens when the fix and
perm may be more appropriate. Does the BD data sheet not have a
recommendation?

Good luck,
Derek