Re: Adjusting staining for large cell numbers

From: Ruud Hulspas <ruud-hulspas@cytonome.com>
Date: Thu Jun 26 2008 - 09:52:20 EDT
Hi Hadi,

When you titrated your antibody using only a million nucleated BM cells 
and found out that 2 ul of antibody in 0.1 ml solution works best, you 
determined the proper antibody concentration. The total amount of 
antibody molecules in 2 ul is generally several fold more than the 
number of available antigens and can therefore also be used to label 
10^8 nucleated BM cells ... as long as you keep the antibody 
concentration the same and therefore the total volume the same.
In your case, it means you need to make a cell suspension in 98 ul 
(~10^9 cells/ml) and add 2 ul of antibody. For confirmation sake, you 
should ask the antibody manufacturer what the antibody concentration in 
the vial is (1 ug/ml ?) and you can do the math to determine how many 
antibody molecules you're using.
I do this all the time and it works great. In fact, using very high cell 
concentrations (and minimum amount of antibody molecules) give much 
better results because it reduces background labeling.
By the way, this topic has been discussed several times on this list 
before. A quick search will give you additional info.

Good luck,

Ruud

-- 