Hi Eric, The simplistic take on it is that compensation is solely dependent on the percentage of spillover of each fluorochrome between channels, and autofluorescence has no bearing on this. Having said that, and it sounds like you know, for initial compensation setup you need to have stained and unstained particles that have identical autoflourescence (see Mario Roederer's compensation site for a good review). Once this is done however, the compensation matrix will be valid for particles of all autofluorescence levels...just be aware that the unstained cells of one type are likely to be in a different position on a plot to unstained cells of another type. Even if you normally do your compensation using cells, BDs comp beads are worth their weight in gold for setting up compensation in populations with heterogeneous AF. If you are working in the mouse, set your viability stain up on normal mouse spleen and try to only get lymphocytes (live and dead) in the scatter gate as they have relatively homogeneous AF. If you have enough free channels in your experiment, think about dedicating one or more solely to autofluorescence. Instead of it being a hindrance it can actually be useful in distinguishing different cell populations, particularly in teasing apart different myeloid lineages. The FITC channel on the calibur is good for this because it can be used essentially without any compensation most of the time, but you can use other channels if you compensate the true fluorochromes out using your preferred flavour of post-acquisition analysis software (Diva is a pain for this but with FlowJo it can be done relatively easily). If you can dedicate 2 (or more) channels to AF, particularly with different excitation lasers (though the red laser on the calibur isn't great), you can readily distinguish the AF+ cells since AF is typically excited over a broad range of wavelengths and emission generally "correlates" on a bivariate plot to form a diagonal "smear" in the 2 channels...though unfortunately you can't really use AF as a "fluorescence parameter" in its own right, probably because it's a result of more than one fluorochrome species with similar, but non identical spectral properties. One last tip...when analysing your data it can be best to initially segregate cells on AF level, if you are using a standard hierarchical gating strategy instead of cluster analysis etc, and only subsequently look at marker expression on cells with similar AF levels. Good luck! Andrew Andrew Mitchell Ph.D Equipe Giovanna Chimini ABCA1 transporters Centre d’Immunology de Marseille-Luminy CNRS-INSERM-Université de la Méditerranée Campus de Luminy, Case 906 13002 Marseille, Cedex 9 France Tel: +33 (0) 4 91 26 94 90 Fax: +33 (0) 4 91 26 94 30 e-mail: mitchell@ciml.univ-mrs.fr ________________________________________ De : Eric Shaw [eseric47@googlemail.com] Date d'envoi : mercredi 11 juin 2008 18:20 À : Cytometry Mailing List Objet : autofluorescence and compensation Hi Flowers, I have a question. I have a tissue dissagregation which will have lots of dead cells and various cell types which most likely have different amounts of autofluorescence. I know I need to use single stained controls (unstained, 7AAD, FITC,PE,APC) but wont my compensation be inaccurate for FITC/PE/APC since I am compensating for the whole population- live and dead? And what about the different cell types (epi's, tissue macrophages etc) that can give differing amounts of autoflorescence? I compensate with my R1 gate off FSC/SSC and I am using a FACS Calibur. All replies will be appreciated. Thanks, Eric ShawReceived on Mon Jun 16 12:38:00 2008
This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST