RE: green laser suitability

From: Telford, William (NIH/NCI) [E] <telfordw@mail.nih.gov>
Date: Wed Apr 30 2008 - 00:23:24 EDT
Hi Ralph...

The 532 nm laser will excite all but the longest of the red FPs (like mPlum) reasonably
well, but is particularly good for the shorter wavelength red proteins, like mOrange,
DsRed and dTomato.  We get better excitation of the longer red FPs (mStrawberry, mCherry,
mPlum, mKate and Katushka) with 561 and 580 nm sources.

532 nm lasers suitable for flow are available at power levels at least up to 150 mW. In
our hands, fluorescent protein s/n sensitivity does increase with higher power levels -
but our 50 mW 532 nm module has certainly been adequate for most detection tasks.  561 nm
lasers are now available up to 100 mW, and our 580 nm laser (MPBC, Canada) peaks at over
200 mW, although we usually run it at 50 mW on our LSR II, and never more than 100 mW.	

Take care,

Bill Telford
NCI-NIH

-----Original Message-----
From: Rossi Ralph [mailto:ralph.rossi@petermac.org]
Sent: Tue 4/29/2008 2:20 AM
To: cyto-inbox
Subject: green laser suitability
Dear all,

	does anyone have any recommendations/experience using ,green, 532nm lasers for
excitation of the various "fruit" (RFP variants: cherry ,tomato, etc) report proteins.
	The maximum power available( that I can find ) for one of these lasers is 50 mw ,
I would think this is adequate, is it  ?

thanks
				ralph

Regards 
Ralph Rossi
Flow Facility Manager
Peter MacCallum Cancer Centre
Melbourne , Australia
Email: ralph.rossi@petermac.org
Ph 96563747 , 96561955



This email (including any attachments) may contain 
confidential and/or legally privileged information and is 
intended only to be read or used by the addressee.  If you 
are not the intended addressee, any use, distribution, 
disclosure or copying of this email is strictly 
prohibited.  
Confidentiality and legal privilege attached to this email 
(including any attachments) are not waived or lost by 
reason of its mistaken delivery to you.
If you have received this email in error, please delete it 
and notify us immediately by telephone or email.  Peter 
MacCallum Cancer Centre provides no guarantee that this 
transmission is free of virus or that it has not been 
intercepted or altered and will not be liable for any delay 
in its receipt.
Received on Wed Apr 30 13:58:00 2008

This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST