RE: Blood platelet sorting using a BD Aria...

From: Ian Dimmick <ian.dimmick@newcastle.ac.uk>
Date: Fri Apr 11 2008 - 16:29:04 EDT
We sort platelets routinely into 96 well plates using a FACSaria. The platelets are in a
standard EDTA anticoagulant bottle , same as is used for routine blood counting . The set
up is to add 1ul of whole blood into 2 ml of sheath fluid , the scatter is set to log
forward and side scatter and threshold is set on FS  200 , this setting can vary
slightly. Initially you will see the majority of events as  red cells to the upper right
of the scatter plot , platelets are situated as a tail below and slightly left of this
red cell population. The gate is fixed to a position so to  avoid the low scatter red
cell events and the low platelet events that may contain some noise and debris. If you
wish you can confirm the purity of this gate by using a CD61 marker however we find that
this is not necessary. 
The set up for our Aria is , 100 micron nozzle , pressure 20 PSI , blue laser at 100mw
.we have not had problems associated with platelet clumping or aggregation within the
tubing , however I can not comment on the activation of the platelets as we do not test
for this , in addition the sorting and anticoagulant used will have to be appropriate to
your downstream application.

Hope this is helpful 


Ian

Ian Dimmick
Flow Cytometry Core Facility Manager
North East England Stem Cell Institute
Bioscience Centre
International Centre for life
Newcastle Upon Tyne
NE1 3BZ
UK
Ian.Dimmick@ncl.ac.uk
Tel 0044 191 2418831
Fax 0044 191 2418666
(mob) 0044 7970344823



-----Original Message-----
From: Benoit Labarthe [mailto:benoitlabarthe@gmail.com]
Sent: Fri 11-Apr-08 3:08
To: cyto-inbox
Subject: Blood platelet sorting using a BD Aria...
Hello Flowers!

We are starting to sort blood platelets using a BD Aria. Our first
experiment was relatively successful for a first short run but does anybody
have experience and advices with this type of sorting? Is there an increased
risk of clotting inside the tubing? Are the platelets partially or totally
activated during the sorting? I guess that the pressure, laser and the
electrostatic charge will not leave platelets unmodified...
Since platelets may also be activated during a long sorting, we are planning
to use high concentrations of platelet inhibitors but, is there a maximum
length to avoid an excessive activation?

Thank you all!

benoit


-- 
Benoît Labarthe, PharmD
Institut de Cardiologie de Montréal
Montreal Heart Institute
Centre de Recherche, Laboratoire du Dr.Théroux
5000 Bélanger Est
Montréal Québec Canada H1T 1C8
Tel (514)-376 3330 ext 3017
Fax (514)- 376 1076
Received on Mon Apr 14 16:18:00 2008

This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST