FW: How to deal with ARIA contamination? another 2 cents worth.

From: Rosson, Dan P <drosson@utmem.edu>
Date: Fri Feb 01 2008 - 09:54:12 EST
I firmly beleive Mike is right! I came to UT about a year ago which was
6 months after they got the instrument. The autoclave cycle for the
sheath fluid wasn't quite enough. Anyway, the first 12 months there were
no contamination problems, then suddenly bugs all over the place! The
filter had eventually developed pathways big enough for the bugs to get
through. 
 
The rest of the story, I believe, is the post by Wayne Green which
alludes to a study that concluded that there are pockets of sheath fluid
that don't get flushed during a cleaning, thus it keeps coming back. The
moral of the story seems to be don't let your instrument get
contaminated in the first place.
 
Alice Givans writes her instrument never had any contamination. I wonder
if she would be willing to deliberately contaminate it to see if she
could get it decontaminated again. ----Just a thought!
 
 
Dan Rosson PhD
University of Tennessee 

________________________________

From: Michael Hughes [mailto:MHughes@picr.man.ac.uk] 
Sent: Friday, February 01, 2008 5:18 AM
To: cyto-inbox
Subject: RE: How to deal with ARIA contamination?


Hi Dan,
      We have been running and Aria for about 3 years. Because we do
sterile sorts every day we really needed to close all the doors to
bacteria. The main problem we found was that the Pall filter on the
sheath line eventually broke down, some quicker than others and so we
replace it with a plastic millipore filter with a 0.22 micron filter we
could replace every day with a freshly autoclaved one. We also replaced
the bubble filter with a stainless steel Millipore filter which again we
replace every day. We use this as backup just in case a filter splits.
 
Regards, Mike Hughes, PICR, Manchester, England 

________________________________

From: Rosson, Dan P [mailto:drosson@utmem.edu] 
Sent: Wednesday, January 30, 2008 7:18 PM
To: cyto-inbox
Subject: RE: How to deal with ARIA contamination?


Join the club! 
 
We developed contamination problems after about 6 months of service in
our Aria. I never had it in a MoFlo or Elite. The "prepare for asceptic
sort" procedure that BD gives you involves removing the filters and then
sterilizing the unit with 70% EtOH. The filters that they give you are
supplied sterile to BD by Pall and Whatman. However, BD puts non-sterile
connections on them and supplies them to you in bubble wrap. If you
autoclave the unit, the water-tight connections are destroyed. I tried
buzzing them in a Cesium irradiator for 10 times longer than the charts
said was necessary, but the damned bugs survived. I never figured out
why. BD says not to use EtOH on the filters especially the bubble
filter, but I chased down its manufacturer and it turns out to be a
Whatman AS 36 which is made of nylon and glass fiber. ( It doesn't take
a PhD in chemistry --which I am-- to tell you EtOH won't hurt it). The
Pall filter is also EtOH resistant (Pall doesn't make a filter that
isn't reisistant to EtOH). So a few weeks back I performed the "prepare
for asceptic sort" and didn't take the filters off. The instrument
didn't dissolve and things were sterile for a while after that.
Recently, however, a couple of tests came back contaminated and the
insidious part of it is it seems to take 4 or 5 days for the
contamination to grow up. Anyway, I'm re doing the sterilization
procedure as I write this.
 
Well my two cents worth in this is  to go ahead and run 70% EtOH through
the entire unit including filters at what ever frequency you think is
appropriate ( I'm thinking every week or two) and do frequent sterility
tests and thus deal with the problem that way. Additionally, I have
users that routinely use gentamycin in their cultures and they never get
hit with contamination. Of course that's just my contamination: yours
may be different.
 
 
Dan Rosson PhD
University of Tennessee

________________________________

From: akos.szilvasi@novartis.com [mailto:akos.szilvasi@novartis.com] 
Sent: Monday, January 28, 2008 1:56 PM
To: cyto-inbox
Subject: How to deal with ARIA contamination?



Dear FLOWers, 

We have a less than two years old (Special Order - UV equipped) ARIA
sorter that had an unrelenting sheath fluid contamination problem. The
recurring mysterious contamination existed from the arrival of the
sorter. We reported it multiple times. BD replaced a few parts of the
fluidics system but it kept the bugs out of the sheath only for 4-6
weeks at the time. Filter replacements, bleaching the system works only
temporarily. Our other Aria sorter, an older one, never had any such
symptoms. We use them in an identical fashion. 

One lab suggested to follow their solution (this is not an isolated
occurrence - not matter what they say) to fill the distilled water tank
with 70% ethanol so that the fluidics shut down procedure would fill the
lines with alcohol. That sounds good but no one could tell if this trick
has some detrimental effect on any parts of the fluidics system. Have
you used alcohol over night in the sorter? Did it cause any problem on
the long run? 

Any other solution? BD offers to replace the whole fluidics system for $
13,000. 

Best regards, 
Akos

__________________________ 

Akos Szilvasi
NIBRI Core Laboratory Services
manager
USCA, 601-5301
Novartis Institutes for BioMedical
Research, Inc.
100 Technology Square
Cambridge, MA 02139
USA
Phone: +1 617 8717177
Email : akos.szilvasi@novartis.com <mailto:akos.szilvasi@novartis.com> 





_________________________

CONFIDENTIALITY NOTICE

The information contained in this e-mail message is intended only for
the exclusive use of the individual or entity named above and may
contain information that is privileged, confidential or exempt from
disclosure under applicable law. If the reader of this message is not
the intended recipient, or the employee or agent responsible for
delivery of the message to the intended recipient, you are hereby
notified that any dissemination, distribution or copying of this
communication is strictly prohibited. If you have received this
communication in error, please notify the sender immediately by e-mail
and delete the material from any computer.  Thank you.

________________________________

This email is confidential and intended solely for the use of the
person(s) ('the intended recipient') to whom it was addressed. Any views
or opinions presented are solely those of the author and do not
necessarily represent those of the Paterson Institute for Cancer
Research or the University of Manchester. It may contain information
that is privileged & confidential within the meaning of applicable law.
Accordingly any dissemination, distribution, copying, or other use of
this message, or any of its contents, by any person other than the
intended recipient may constitute a breach of civil or criminal law and
is strictly prohibited. If you are NOT the intended recipient please
contact the sender and dispose of this e-mail as soon as possible. 



ATT15553549.gif
Received on Fri Feb 1 13:18:00 2008

This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST