Re: mCherry on a Calibur ??!!

From: Adrian Smith <a.smith@centenary.usyd.edu.AU>
Date: Thu Aug 30 2007 - 19:36:32 EDT
On 29/08/2007, at 11:50 PM, Rosemary Clarke wrote:

> Dear All,
>
> Now this is going to sound like a very stupid question, so  
> apologies in advance.
>
> Is it possible to detect mCherry fluorescence on a FACS Calibur,  
> using 488nm excitation and detecting fluorescence at 585nm? I have  
> always thought not as (in theory at least) mCherry is not excited  
> at 488nm. However, I have just had a user come to me who thinks he  
> may be getting a good signal (two log shift from untransfected  
> cells) with what they believe to be mCherry. However, there could  
> have been a mix up and it may be dsRed (don't ask!), which would  
> obviously be detectable. They are in the process of checking and  
> sequencing the constructs but I am now paranoid that I have got it  
> wrong and mCherry CAN be detected. Has any one out there in flow- 
> land managed to get a signal from mCherry on a Calibur using 488nm  
> excitation?
>
> Thanks in advance
> Rosie
>

Hi Rosie

In testing mCherry on our VantageDiva we can see some signal from the  
488 (@200mW)... but it is poor, especially compared to excitation by  
the dye laser (or to GFP). Two logs on the Calibur seems unlikely,  
particularly as I think the dectection filter in FL2 on Calibur is  
also sub-optimal.

Note that in attached plot we had to turn the O630_6 voltage way down  
(about 300V I think rather than our normal 600), ie this was a very  
bright sample...

Regards,

Adrian