(no subject)

From: <owner-cytometry>
Date: Fri Aug 17 2007 - 12:26:18 EDT
From mailnull@flowcyt.cyto.purdue.edu Thu Aug 16 15:42:46 2007
Received: from flowcyt.cyto.purdue.edu (localhost.cyto.purdue.edu [127.0.0.1])
	by flowcyt.cyto.purdue.edu (8.12.10/8.12.10) with ESMTP id l7GJgg5A003380
	for <cytolist@flowcyt.cyto.purdue.edu>; Thu, 16 Aug 2007 15:42:42 -0400 (EDT)
	(envelope-from mailnull@flowcyt.cyto.purdue.edu)
Received: (from mailnull@localhost)
	by flowcyt.cyto.purdue.edu (8.12.10/8.12.10/Submit) id l7GJggtR003379
	for cytolist; Thu, 16 Aug 2007 15:42:42 -0400 (EDT)
	(envelope-from mailnull)
Received: from mail2.qmul.ac.uk (mail2.qmul.ac.uk [138.37.6.6])
	by flowcyt.cyto.purdue.edu (8.12.10/8.12.10) with ESMTP id l7GJgb2a003373
	for <cytometry@flowcyt.cyto.purdue.edu>; Thu, 16 Aug 2007 15:42:38 -0400 (EDT)
	(envelope-from U.Johansson@qmul.ac.uk)
Received: from termite.qmul.ac.uk ([138.37.6.52])
	by mail2.qmul.ac.uk with esmtp (Exim 4.66)
	(envelope-from <U.Johansson@qmul.ac.uk>)
	id 1ILlEn-0001g1-FA
	for cytometry@flowcyt.cyto.purdue.edu; Thu, 16 Aug 2007 20:42:35 +0100
Received: from localhost.localdomain (termite [127.0.0.1])
	by termite.qmul.ac.uk (8.13.8/8.13.8) with ESMTP id l7GJgXM3019318
	for <cytometry@flowcyt.cyto.purdue.edu>; Thu, 16 Aug 2007 20:42:33 +0100
Received: (from apache@localhost)
	by localhost.localdomain (8.13.8/8.13.8/Submit) id l7GJgXIp019315
	for cytometry@flowcyt.cyto.purdue.edu; Thu, 16 Aug 2007 20:42:33 +0100
Received: from 88-109-135-179.dynamic.dsl.as9105.com
	(88-109-135-179.dynamic.dsl.as9105.com [88.109.135.179]) by
	webmail.qmul.ac.uk (Horde MIME library) with HTTP; Thu, 16 Aug 2007
	20:42:33 +0100
Message-ID: <20070816204233.u7oym7kk08ks8gog@webmail.qmul.ac.uk>
Date: Thu, 16 Aug 2007 20:42:33 +0100
From: Ulrika Johansson <U.Johansson@qmul.ac.uk>
Subject: Re: CFSE compensation
References: <007101c7ddcf$ed2a9c70$3c6fa8c0@Candela>
	<8C9AD22D6F53CBD-ED4-414F@FWM-D15.sysops.aol.com>
	<8C9AD230E291635-ED4-4158@FWM-D15.sysops.aol.com>
In-Reply-To: <8C9AD230E291635-ED4-4158@FWM-D15.sysops.aol.com>
MIME-Version: 1.0
Content-Type: text/plain;
	charset=UTF-8;
	DelSp="Yes";
	format="flowed"
Content-Disposition: inline
User-Agent: Internet Messaging Program (IMP) H3 (4.1.4)
X-Originating-IP: 88.109.135.179
X-QM-Sender: hiw070@imap.qmul.ac.uk
X-Sender-Host-Address: 138.37.6.52
X-QM-Scan-Virus: ClamAV says the message is clean
X-QM-Scan-Virus: virusscan says the message is clean
X-Spam-Score: -4.9 () BAYES_00
X-Scanned-By: MIMEDefang 2.39
Content-Transfer-Encoding: 8bit
X-MIME-Autoconverted: from quoted-printable to 8bit by flowcyt.cyto.purdue.edu id l7GJgg2a003375
X-PMFLAGS: 34078848 0 1 3385.cnm                            

Hi Maria,

You mention that the FL1 peak is not very narrow; is this for the  
stimulated T cells? I too wonder if your customer has observed	
proliferation (if a mitogen is used it should be obvious just lookig  
down the microscope).

Just a thought, but if a high concentration of CSFE is used then I  
know from experience that it can sometimes be difficult to discern  
peaks; and if a mitogen has been used, and if your researcher is  
homing in on T cells, it's likely that all T cells have proliferated  
(all this may thus result in a broad looking sort of 'peak'.

Good luck,
BW
Ulrika


Quoting rozenkov@netscape.net:

>
> Hi Maria,
>
> It is rather difficult to make CFSE-T cells non-proliferating by   
> compensation. How does the single-parameter FL1 (FITC) histogram   
> look? Are all cells contained in the same peak as in a   
> non-stimulated control? (Seeing the pictures here would help.)
>
> I guess even if you have done 6-colour auto-comp, you can just run a	
>  CFSE-only (no antibodies) stimulated cell sample and see how it  
> looks.
>
> Another question: does you colleague see proliferation in the same   
> samples by other parameters, e.g. thymidine incorporation or at   
> least morphologically?
>
> Regards.
>
> Vladislav Rozenkov
>
>
> -----Original Message-----
> From: Maria Candela Iglesias Chiesa <candela@iner.gob.mx>
> To: cyto-inbox
> Sent: Tue, 14 Aug 2007 3:32 am
> Subject: CFSE compensation
>
>
>
>
> Hi everybody,
>
>  
>
> A colleague is putting up a CFSE proliferation assay on T cells. She	
>  hasn´t been able to see any proliferation after 6d of stimulation.	 
> However, I strongly suspect this is a problem of compensation more   
> than anything else. We have a FACSAria and normally do Compbeads and	
>  automatic compensation. She has run non stimulated cells CFSE   
> stained to compensate for the “FITC” channel, and stained compbeads   
> for the other channels. She hasn’t got a very narrow peak in “FITC”,  
>  but the compensation works anyway. Still, she doesn’t see any   
> proliferation, and has performed her experiment in a very standard   
> way. Am I missing some obvious point in the compensation procedure?	
> How do you normally compensate when using CFSE? I’ve done it   
> manually on a 3 color expt in a Calibur and never had any problem.   
> Should I try and compensate manually even if she is using 6 colors?
>
> Many thanks,
>
>  
>
> Candela
>
>  
>
> María Candela Iglesias, PhD
>
> Centro de Investigacion en Enfermedades Infecciosas (CIENI)
>
> Instituto Nacional de Enfermedades Respiratorias
>
> Calzada de Tlalpan 4502. Col. Seccion XVI
>
> CP 14080. Deleg. Tlalpan
>
> México DF.
>
> tel y fax +52 (55) 56 66 79 85
>
> candela@iner.gob.mx
>
> candela.iglesias@cieni.org.mx
>
>  
>
>
>
> Check Out the new free AIM(R) Mail -- Unlimited storage and	
> industry-leading spam and email virus protection.
>
>
>
> ________________________________________________________________________
> Check Out the new free AIM(R) Mail -- Unlimited storage and	
> industry-leading spam and email virus protection.
>



_______________________________
Ulrika Johansson, PhD
Department of Haematology
Flow Suite
Pathology and Pharmacy Building
80 Newark Street
London
E1 2ES
UK

P +44 0203 246 0228
F +44 0203 246 0225
e U.Johansson@qmul.ac.uk
Received on Fri Aug 17 15:18:00 2007

This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST