Is there any studies on the relationship between a size (diameter) of the bead and CV ? Aki On Mar 1, 2007, at 2:50 PM, Nebe-Von-Caron, G wrote: > The correct counting formula when using reference beads for > enumeration is shown below. > > Sample concentration = {[ref volume] x [sample predilution] x > [sample count]x [ref concentration ml^-1]}/{ [sample volume] x [ref > predilution] x [ref count] } > > it does not matter if you dilute your sample in 1 or 50ml as this > only alters the flow rate of the sample, but not the ratios of > absolute counts. > > > If you add equal volumes of reference beads and sample and nothing > is prediluted the formula simplifies to > > Sample concentration ml^-1= {[sample count]x [ref concentration > ml^-1]}/{[ref count]} > > > The Trucount tube delivers a fixed number of beads trapped in a > type of sugar matrix. If you add Xml of sample, your fixed number > of beads are therefore diluted in Xml (or else) so you only need to > ratio: > > Sample concentration ml^-1= [sample count]x [total ref bead > number]}/{[ref count] x [Xml]} > > > A similar simplification can be made if your bead volume is much > smaller than the sample volume. If 10ul beads are added to 1ml of > sample your total volume is 1010ul which is a 1% volume error in > the calculation. Similarly the beads in the true count bead > introduce a small volume error which can be corrected for by > altering the "reference bead figure" so you don't have to do it in > the calculation > > There was some discussion with regards to the errors of those > methods several years ago on the list. The CV's of all pipetting > steps are additive (assume 1% per pipetting step, hoping they are > well maintained, fitting tips and good pipetting practice) and the > counting errors are also additive (square root of count over > count). So if you count 8000 beads but only 100 cells of interest > your theoretical counting CV is 11% (1% in the 5000 and 10% in the > 100). Other problems come from bead dispersion (we correct for > aggregates up to 6 in our aggregation studies) as well as loss of > beads or adhesion to cells, tube walls .. > > > Whilst in theory all counting methods should count the same, there > seems to be that well known discrepancy between theory and practice. > > http://www.affordcd4.com/pdf/Affref07_BB.PDF > > good luck > > > Gerhard > > > > In the case of the > > > > -----Original Message----- > From: donny thomas [mailto:donny@reametrix.com] > Sent: 28 February 2007 13:14 > To: Cytometry Mailing List > Subject: Re: Calculation of Absolute Count > > Sunny > > We use the second method you tried, and it works fine with us. > > Abslute count = No: of gated cells/No: of gated beads X Total bead > count per test(tube)/test volume > > Did you try the above mentioned formulae with the Coulter beads? Do > you know the no: of beads you use in one test. ?(eg: 50,000 beads > per 50uL of blood) > > BD FACScount tubes come with presice bead counts per tube. > > Donny > > On 2/27/07, Xiao Qun Zhai <xqzhai@uclan.ac.uk > wrote: > Dear All, > > I use Absolute Count beads from Beckman Coulter - Product No. 7547053. > > On the data sheet, the formula for Absolute Count (cells/uL) is: > > Total Number of Cells Counted > ------------------------------------- x *Beads Concentration (which > is known from the datasheet). > Total Number of Beads counted > > Beckman Coulter's technician said I don't need to divide this > result by > dilution factor. If so, I would think the *Beads concentration here > should be the bead conc. in the fianl FACS tube not in the original > bottle. > > However, BD's specilist gave me another formula (here we have BD > FACSAria), which is > > Absolute count = No. of cells in Gate & nbsp; Bead > count (on > packet) > ; -------------------------&nb sp; X > ------------------------------- > &n bsp; & nbsp; No. of beads in > gate & nbsp; Volume of > FACS tube sample > > Which one is right? I used both of them but neither of the result is > accordance with my haemacytometer result. > > I am confused. > > Thanks for help. > > Sunny > >Received on Mon Mar 5 11:58:00 2007
This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST