Before you set the sweet spot, you need to compare the actual pixel number
of drop 1 (on the right) to the target pixel number (on the left.)  If these
numbers aren't close you need to change the target number to match the
actual number.	If you don't do this, the sweet spot has a hard time
determining where you want that break and will jump back and forth between
the two breakoff points.


Pam Shaw

Fluorescence Cytometry Core Facilitator

Center for Environmental Health Sciences

The University of Montana

32 Campus Dr, Skaggs Building 052

Missoula, MT 59812


Ph (406) 243-4974

email:	 <mailto:pamela.shaw@umontana.edu> pamela.shaw@umontana.edu

www.umt.edu/cehs <http://www.umt.edu/cehs/FluorCytomCore.htm>
/FluorCytomCore.htm


-----Original Message-----
From: Furgeson, Alice N [mailto:Alice.N.Furgeson@pfizer.com] 
Sent: Wednesday, December 06, 2006 10:17 AM
To: cyto-inbox
Subject: FACSAria unstable drop 1 and gap


Hello, 
I was wondering if any Aria users have experienced unstable drop 1 and gap
even with the Sweet Spot on.  I use a 100um nozzle.  I have de-bubbled the
filters, cleaned the flow cell with contrad 70, sonicated the nozzle and
changed the O-ring.  I have also tried using a different 100um nozzle and a
70um nozzle.  The gap still fluctuates between 2 and 20 and the drop 1 goes
between 100 to 300.  The numbers just bounce everywhere.  Any suggestion for
why this is happening and any solution?