RE: FACSAria unstable drop 1 and gap

From: Hirschkorn, Dale <dhirschkorn@bloodsystems.org>
Date: Thu Dec 07 2006 - 13:47:23 EST
Hello Alice,

 

I had the same problem with my Aria.  It turned out that I needed a new
transducer-talk to you field service engineer about getting a new one.
I have not had the problem since.  Hope you have a service contract!

 

Dale Hirschkorn

Blood Systems Research Institute

270 Masonic Ave.

San Francisco, CA  94118

(415) 749-6672

dhirschkorn@bloodsystems.org

www.bloodsystems.org <http://www.bloodsystems.org/> 

 

-----Original Message-----
From: Furgeson, Alice N [mailto:Alice.N.Furgeson@pfizer.com] 
Sent: Wednesday, December 06, 2006 9:17 AM
To: cyto-inbox
Subject: FACSAria unstable drop 1 and gap

 

Hello, 
I was wondering if any Aria users have experienced unstable drop 1 and
gap even with the Sweet Spot on.  I use a 100um nozzle.  I have
de-bubbled the filters, cleaned the flow cell with contrad 70, sonicated
the nozzle and changed the O-ring.  I have also tried using a different
100um nozzle and a 70um nozzle.  The gap still fluctuates between 2 and
20 and the drop 1 goes between 100 to 300.  The numbers just bounce
everywhere.  Any suggestion for why this is happening and any solution?

Thanks, 
Alice 
Received on Fri Dec 8 13:18:00 2006