Alice: Have you played with either the sheath pressure or the frequency? At low pressure, I've found I can get a very stable stream at 25 PSI, and a frequency of 32. At high pressure, I with a frequency of 92-94. Have you examined the nozzles with a microscope to make sure that everything is fine, no pitting or corrosion. Regards Tim ________________________________ From: Furgeson, Alice N [mailto:Alice.N.Furgeson@pfizer.com] Sent: Wednesday, December 06, 2006 12:17 PM To: cyto-inbox Subject: FACSAria unstable drop 1 and gap Hello, I was wondering if any Aria users have experienced unstable drop 1 and gap even with the Sweet Spot on. I use a 100um nozzle. I have de-bubbled the filters, cleaned the flow cell with contrad 70, sonicated the nozzle and changed the O-ring. I have also tried using a different 100um nozzle and a 70um nozzle. The gap still fluctuates between 2 and 20 and the drop 1 goes between 100 to 300. The numbers just bounce everywhere. Any suggestion for why this is happening and any solution? Thanks, AliceReceived on Fri Dec 8 12:58:00 2006