RE: receptor internalisation analysis

From: David Basiji <basiji@amnis.com>
Date: Wed Sep 13 2006 - 16:00:42 EDT
Hi Guy,
 
Since you asked...
 
You're correct that the ImageStream system is configured like a
traditional widefield microscope, not a confocal scope. Our 0.75 NA
objective results in an optical depth of field of about 1 micron. This,
combined with hydrodynamic focussing of the cells with a positional
accuracy of ~1.5 microns, results in a very consistent equatorial
"coffee ring" image of membrane-bound fluorescence signals. 
 
I would say that the change from a sharp "coffee ring" staining pattern
to one that is uniformly distributed over a relatively wide band of
cytoplasm is pretty good evidence of internalization, and in my opinion
it can be done well with our system. However, I would also agree that
any methodology relying on an uncontrolled assessment of fluorescence
distribution alone would benefit from a decreased depth of field like
that provided by confocal imaging. 
 
In my opinion, it's preferable to use an internal control for surface
localization. The way we generally assess internalization is to label
each cell with a known surface-bound control marker in one color along
with the experimental marker which is being assessed for internalization
in a second color. We then perform a cross-correlation of the two images
of each cell and look for a decrease in image correlation as evidence
that the experimental marker is redistributing. This methodology detects
capping and/or internalization because any relative signal
redistribution between the two images that exceeds 0.5 microns (our
pixel size) is reliably detected. It won't replace FRET if you need
Angstrom or nanometer-scale co-localization (like detecting the inner
versus outer leaflet membrane positioning), but by the same token, it is
very easy to implement and has a much wider "spatial dynamic range" than
FRET. The cross-correlation methodology also be used to determine
compartmentalization if you label the endosomes, lysosomes, or golgi.
Nuclear translocation of transcription factors is assessed by labeling
the nucleus with a DNA binding dye and cross-correlating the nuclear
image to the image of the transcription factor. 
 
There are a number of studies published and in press that utilize this
methodology. If you're interested, I would recommend the following for a
more detailed description of the cross-correlation approach itself:
 
http://www.ncbi.nlm.nih.gov/entrez/query.fcgi?cmd=Retrieve&db=pubmed&dop
t=Abstract&list_uids=16563425&query_hl=1&itool=pubmed_docsum
 
Best Regards,
David
 
 
David Basiji, Ph.D.
CEO, Amnis Corporation
2505 Third Ave., Suite 210
Seattle, WA 98121
(206) 374-7165 direct
(206) 919-3342 mobile
(206) 576-6895 fax
www.amnis.com
 


________________________________

	From: Guy Hermans [mailto:Guy.Hermans@ablynx.com] 
	Sent: Wednesday, September 13, 2006 2:43 AM
	To: Cytometry Mailing List
	Subject: RE: receptor internalisation analysis
	
	
	Hi V.,
	 
	could you please elaborate on that?
	 
	This is the type of discussion I continuously see when
critically evaluating data re: cell membrane binding versus diffuse
internalization. True sceptics will dismiss classical fluorescence
microscopy images showing cytoplasmic staining (diffuse or punctate),
arguing you really need confocal to decide whether you're seeing surface
stain versus intracellular - and I do see their point.
	The way I understand the Amnis device works, confocal is really
not an option. Seeing the individual cell's staining pattern may
therefore be nice to have above and beyond the dotplots/histo's and
whatnot outputs as from a conventional cytometer, but shouldn't be used
as a definitive argument for internalization.
	 
	I would therefore argue you can use the Amnis to gather more
data from your cells, but for this particular application having the
macrofocal image shouldn't make the decisive difference. Doing the
"stripped cells" experiment, using fluor-specific secundaries, quenching
and/or pH sensitive dyes will therefore be as necessary as when using a
non-imaging device.
	 
	Just my two cents, and purely based on my limited knowledge of
the Amnis of course. If it does do confocal as well - hey guys, if you
read this, write it in block capitals on all your flyers and put me down
for one machine to start with!	;-)
	 
	Guy
	 
	 Next generation therapeutic antibodies <http://www.ablynx.com/>

	
Guy Hermans, PhD
Senior Scientist	Ablynx NV
Technologiepark 4
B-9052 Zwijnaarde
Belgium 	
guy.hermans@ablynx.com	tel: 
fax: 
mobile: 	+32 (0)9 261 06 57
+32 (0)9 261 06 27
+32 (0)486 788 551 	
	
Add me to your address book...
<https://www.plaxo.com/add_me?u=30065269879&v0=994426&k0=2009290972>
Want a signature like this? <http://www.plaxo.com/signature> 	

		-----Original Message-----
		From: rozenkov@netscape.net
[mailto:rozenkov@netscape.net] 
		Sent: Tuesday, September 12, 2006 7:12 AM
		To: Cytometry Mailing List
		Subject: Re: receptor internalisation analysis
		
		
		An ImageStream flow cytometer from Amnis could do this
with no additional procedures...
		 
		Regards.
		 
		Vladislav Rozenkov

		-----Original Message-----
		From: julie.bertout@ibl.fr
		To: cyto-inbox Sent: Fri, 8 Sep 2006 9:56 PM
		Subject: receptor internalisation analysis
		
		
		Hi, 
		 
		I have a researcher interested in quantitating receptor
internalisation after different treatments. 
		 
		Here is the way she used to do it : 
		- label receptors of interest with primary antibody and
secondary antibody 
		- do the treatment she wants to test 
		- remove label from non-internalised receptor with a
slightly acid treatment (so only internalised receptors will still be
fluorescent) 
		- analyse her samples fluorescence to see a difference. 
		 
		the problem is that, as antibody/receptor afinity is
different from one receptor to another, she can't do her positive and
negative controls... for example, the receptor which she used as control
(which internalisation shouldn't change with/without treatment) has such
a high affinity with its antibody that she can't remove it with the acid
treatment. 
		 
		So I would like to know if there is any way to quench
fluorescence for the receptor that are not internalised or if anybody
has a protocol to quantitate internalised receptor? 
		I thought of comparing non-permeabilised vs
permeabilised samples but PFA fixation permeabilise cells a little,
doesn't it? 
		 
		Thank you for your answers, 
		 
		Julie Bertout 
		cytometry lab 
		Institut Pasteur de Lille 
		1 rue du professeur Calmette 
		59800 Lille 
		France 
		 
		 
		
________________________________

		Check Out the new free AIM(R) Mail
<http://pr.atwola.com/promoclk/100122638x1081283466x1074645346/aol?redir
=http%3A%2F%2Fwww%2Eaim%2Ecom%2Ffun%2Fmail%2F>  -- 2 GB of storage and
industry-leading spam and email virus protection.
		


________________________________

	This e-mail message is intended only for the use of the
individual or entity to which it is addressed and may contain
information that is privileged, confidential and exempt from disclosure.
If the reader of this E-mail message is not the intended recipient, you
are hereby notified that any dissemination, distribution or copying of
this communication is strictly prohibited. If you have received this
communication in error, please notify us immediately at
ablynx@ablynx.com. Thank you for your co-operation.
	
	"NANOBODY" and "NANOCLONE" are registered trademarks of Ablynx
N.V. 
	
________________________________






Ablynx_logo_mini.gif
Received on Thu Sep 14 14:42:03 2006

This archive was generated by hypermail 2.1.8 : Sat Sep 16 2006 - 03:12:06 EDT