Paul Waring wrote:

> I have been out of touch with the mitochondria field for a few  
> years. Would anyone care to give me an update on the relative  
> advantages of JC-1 versus Dioc6 (or other dyes) in measuring	
> mitochondrial membrane potential ?

JC-1 has been found useful for discriminating energized from  
deenergized mitochondria, but, as far as I know, has never been  
demonstrated to *measure* mitochondrial potential. The distribution  
of green vs. red fluorescence is too broad in all directions to make  
a ratiometric procedure feasible.

It is claimed that DiOC6(3) at very low concentrations (around 1 nM)  
*can* be used for quantitative measurements (Rottenberg H, Wu S:  
Quantitative assay by flow cytometry of the mitochondrial membrane  
potential in intact cells. Biochim Biophys Acta 1404:393-404, 1998).  
I suspect that other cyanines would also work at this low  
concentration. If the mitochondria are energized,  the dye  
concentration inside will be at least 100 times the external  
concentration; if the external concentration is much higher than 1  
nM, there will be substantial quenching of intramitochondrial dye  
fluorescence, and also more interference from dye in the cytosol. At  
the lower concentrations, mitochondrial fluorescence would be  
expected to track membrane potential.

-Howard