The DCs could very well be getting activated by LPS present in your 
system. Macrophages can also be sensitive to the presence of LPS. 
There are microtiter plate based tests that you can use to measure 
LPS levels in your system (we use the LAL QC1000 test from Cambrex).

We have been monitoring LPS levels on our MoFlo for over 2 years now 
and have found the in-line sheath filter to be the greatest source of 
LPS contamination. For prevention, we take care to dump out whatever 
is remaining in the sheath tank and rinse it with 70% EtOH, as well 
as flush the lines with 70% EtOH every evening. We also periodically 
"soak" the inside of the sheath tank with 5 N NaOH to destroy any LPS 
present there and change the sample line in addition to the in-line 
sheath filter. If you need any more details you can contact me off-
line.

Good luck,

Marie Follo


Am 5 Apr 2006 um 16:40 hat Barry Moran geschrieben:

Hi there-
We're sorting cultured murine dendritic cells on a MoFlo (70micron
nozzle, 60psi) into culture media. We're finding that the cells 
appear
some-what activated (CD40+, CD80+, CD86+) after the sort compared to
unsorted controls. Have many others found this, and if so are there
any suggestions or fixes- is it likely to be endotoxins etc in the
sheath fluid/lines? Many thanks Barry. Barry Moran Flow Cytometry
Facility School of Biochemistry and Immunology 0.19, Biotechnology
Building Trinity College Dublin +353 1 608 3575 barry.moran@tcd.ie
 <http://www.tcd.ie/biochemistry/flow> www.tcd.ie/biochemistry/flow


******************************************************************

Marie Follo, Ph.D.
Core Facility
Zentrale Klinische Forschung
Breisacherstr. 66
Universitaetsklinik Freiburg
79106 Freiburg
Germany
Tel.: +49 761 270 6377
e-mail: marie.follo@uniklinik-freiburg.de
http://www.ukl.uni-freiburg.de/core-facility
Received on Fri Apr 7 15:58:00 2006

This archive was generated by hypermail 2.1.8 : Sat Apr 08 2006 - 04:12:02 EDT