Hi Alice, laser light "bounce", otherwise known as diffraction, does occur at the sheath-sample interface when running liquids with different refractive indices such as ddH2O, bleach, etc.. This is readily detected in "forward" scatter (FSC) with log-amplification as one may use to detect small particles such as bacteria. It is not uncommon to attribute the noise from this diffraction to debris in the sample. good luck, TomD Alice L. Givan wrote: >To Howard et al, >As Howard said, you cannot use distilled water in the sheath stream if you are sorting as >distilled water will not take the charge needed for sorting. > >But, even for non-sorting applications, I have always thought (although have not tested) >that using distilled water as sheath fluid will raise background signals --- because >light will bounce off the interface between the core stream and the sheath fluid (they >no longer have the same refractive index). > >Is this incorrect? > >Alice > >Alice L. Givan, Director >Englert Cell Analysis Laboratory >of the Norris Cotton Cancer Center >Dartmouth Medical School >Lebanon, NH 03756 USA >tel 603-650-7661 >fax 603-650-6130 >givan@dartmouth.edu >www.dartmouth.edu/~celllab > > > >Received on Mon Feb 20 14:18:00 2006