Alice Givan wrote: >To Howard et al, >As Howard said, you cannot use distilled water in the sheath stream >if you are sorting as >distilled water will not take the charge needed for sorting. > >But, even for non-sorting applications, I have always thought >(although have not tested) >that using distilled water as sheath fluid will raise background >signals --- because >light will bounce off the interface between the core stream and the >sheath fluid (they >no longer have the same refractive index). Alice is right about light being scattered at the boundary of the core and sheath streams if the refractive indices are different, as is the case when a distilled water sheath is used for analysis of a sample in a saline solution. The background of *scatter* signals is then increased, although, at least in my experience, this becomes a problem only when one analyzes particles with very small scatter signals, such as bacteria. As was suggested in my earlier posting on this subject, I can usually get away with using a distilled water sheath even for analyses of bacteria. The Technicon/Bayer hematology instruments, which measure both scatter and absorption (highly sensitive to index mismatch), benefit from the use of index-matched sheaths, or at least the older ones did. I suspect that any difference in fluorescence background due to core-sheath refractive index differences is small; I can get 1-2% CVs in fluorescence measurements of dyed beads or Hoechst dye-stained nuclei without matching refractive indices of core and sheath. -HowardReceived on Mon Feb 20 12:38:00 2006