RE: Calibrating Across Continents

From: ALT Service <altservice@direcway.com>
Date: Tue Jul 12 2005 - 11:34:06 EST
To All,
I feel this topic very important.

This reply is not intended as a solicitation, rather to share and check the
experience I have gained.

I agree with Robert on most points totally. In my experience, instrument
stability is a direct result of instrument maintenance.  A higher level of
maintenance will give a more stable instrument. Like Robert, I suspect from
the operation description given by Russ, many instruments are not operating
at optimum levels and a few are probably not even operating at acceptable
levels, but what can you say, FACScomp passed anyway.

In my opinion, A good QA procedure should include both an instrument control
and an assay control as well as concerns for flow rates and CV's at a
minimum.

Concerning the issue of calibration. According to GMP(Good Manufacturing
Practices), when using the term calibration in the US, one should understand
the implication that a standard exists which is traceable to NIST. Such is
the case with size, weight or electronic standards, etc.. For example, when
we have our test equipment, scales & thermometers, etc. calibrated they are
compared to a standard which is traceable to NIST. A report and a sticker
are provided to indicate and verify such traceability. To my knowledge, no
traceable standard exists for fluorescence emissions. MESF values can be
helpful, but not traceable to NIST.

>From a regulatory stand point, how can we calibrate a flow cytometer when no
traceable standard exists to compare readings to?
Technically, I believe we can't.

Perhaps "Correlation" would be a more accurate description.

Questions / comments invited.
Best regards,
Tony Leger
Automation Lab Technology
altservice@direcway.com
1-360-983-8690

--