RE: Looking for advice on pediatric samples

From: Guy Hermans <Guy.Hermans@ablynx.com>
Date: Thu Jun 23 2005 - 02:27:44 EST
Hi Bunny,
 
I'd suggest either setting up the instrument using a large sample available,
stained and with controls identical to your actual sample and running your
actual sample afterwards - no need to waste cells optimizing settings that
way.
 
Other alternative is to take your sample to a scanning microscope type
cytometer; no single cell should get lost that way.
 
I've got a FACSarray, but can't see a big advantage over a conventional
cytometer in terms of number of cells required for analysis. It does use a
syringe drive type of injection mechanism, so maybe a little less volume
gets wasted - but you could fix that up for your sample simply by diluting
it a bit more than usual anyway.
 
Guy
 



	 Next generation therapeutic  <http://www.ablynx.com/> antibodies

	

Guy Hermans, PhD
Project Leader	Ablynx NV
Technologiepark 4
B-9052 Zwijnaarde
Belgium 	
guy.hermans@ablynx.com 	
tel: 
fax: 
mobile: 	+32 (0)9 261 06 57
+32 (0)9 261 06 27
+32 (0)486 788 551 	
	

 <https://www.plaxo.com/add_me?u=30065269879&v0=994426&k0=2009290972> Add me
to your address book...   <http://www.plaxo.com/signature> Want a signature
like this?	

-----Original Message-----
From: bunny [mailto:bunny@cotleur.com] 
Sent: Tuesday, June 21, 2005 11:14 PM
To: cyto-inbox
Subject: Looking for advice on pediatric samples


Hello Flowers!

I'm looking for advice from someone/anyone who is currently analyzing small
cell numbers by flow cytometry. (NOT low frequency, but very few events to
begin with).

We are planning to start analyzing pediatric CSF samples in the near future.
Since we are expecting very SMALL sample volumes with very FEW cells, I'm
looking for advice on the best way to simultaneously quantify chemokine
receptors, cytokine expression and cell phenotypes. 

Is flow cytometry the best method in this case? With very few TOTAL cells,
should we try a microscopy approach? Can someone point (or SHOVE) me towards
the best method?

Our current equipment choices:
At the moment, our Flow core is only equipped with a Facscan and a LSR1. I
could feasibly do 4 color (with Viability on a 5th detector, using UV). 
We are waiting to find out if we will be funded for an Aria. In any case, we
could actually take samples across town to be run on an Aria. (I'm thinking
in terms of maximizing # colors at this point).
We will also have a small amt of equipment money- enough for a
FacsArray-type cytometer. Is there an advantage (as far as running smaller
samples) vs the traditional continuous stream? (We've removed the droplet
containment on the LSR1 years ago, so we already fixed the problem with the
DCM sucking the sample before we get the bar pulled over. I regularly run
100ul samples without a problem). 
Our institution also has a well equipped Confocal /Fluorescent Microscopy
core. 

Thanks for any input!

bunny




Bunny Cotleur, M.S.
Sr Research Technologist, Dept of Neurosciences
Scientific Consultant, Flow Cytometry Core