RE: The Mystery of The High Throughput Sampler, The Sampler Port, and the Clumped Cells

From: <akos.szilvasi@novartis.com>
Date: Fri May 20 2005 - 14:59:35 EST
Unfortunately the HTS software (now part of Diva) on the LSR II does not 
allow you to delay the acquisition. That's a (would be) a very important 
feature for many reasons and it is available on the Macintosh for the 
Calibur. The reason of the disturbance is too high cell concentration 
and/or presence of many RBCs. It is a hydrodynamic disturbance that 
settles after a couple of seconds of run. In the absence of acquisition 
delay you should dilute your samples and try to eliminate RBSs as much as 
possible. We hope BD is polishing up the HTS software on Diva. Especially 
because I am about to order three of them.

Akos



________________________________________________

Akos Szilvasi

Flow Cytometry and Confocal Microscopy Core Facilities
Novartis Institutes for Biomedical Research, Inc. (NIBRI)
100 Technology Square, Cambridge, MA 02139
akos.szilvasi@novartis.com
617 871 7177 office
617 871 7051 fax





"Poon, CK" <cpoon@medsfgh.ucsf.edu>
05/19/2005 05:22 PM

	To:	Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
	cc:	(bcc: Akos Szilvasi/PH/Novartis)
	Subject:	RE: The Mystery of The High Throughput Sampler, The Sampler Port,
and the 
Clumped Cells


Hello,

We use our HTS with the LSRII here at UCSF. 

We have done about 15 runs (PBMCs) on our HTS and have not experienced any 
clogging.
However, I did notice that with our concentration of cells ( ~8million 
PBMC/ml), the dots
start to jump around on the plots if we run faster than 1.5ul/s (a sign of 
clogging?). I
also noticed that keeping the TUBE/PLATE rocker switch in the TUBE postion 
when running
the HTS tends to screw the run up. 

I don't know if this will help and I don't know anything about the boost 
pressure.

Good Luck,
Ck Poon

Research Associate
Core Immunology Lab
UCSF/SFGH


-----Original Message-----
From: Villasmil, Rafael [mailto:VillasmilR@MedImmune.com]
Sent: Wed 05/18/2005 02:32 PM
To: cyto-inbox
Subject: The Mystery of The High Throughput Sampler, The Sampler Port, and 
the Clumped
Cells
Good Afternoon Everyone
I want to find out why the sample port clogs when I use the HTS but it 
works fine when
doing Manual Tube Loading (MTL) 
I have an assay that runs perfectly when I use tubes to load the samples. 
However, the
Calibur sample port clogs constantly when I use the HTS. 
Looking at the samples under the microscope, we found that there are some 
clumped cells. 
We think that the boost pressure of the HTS is higher than the boost 
pressure of the
Calibur when doing MTL
Any comments are welcome
Thanks

Rafael Villasmil 
Associate Scientist I

Medimmune Inc 
Phone Number (301) 398-4478
One MedImmune Way
Gaithersburg, MD 20878 