Hi Ray, First thing I would do is align the laser though the forward scatter channel, when you look at the laser beam image is it a single round spot? How are your signals on beads? Its probably just a reflection off something maybe the window though which the forward scatter signal goes. Try covering that with a bit of paper or tape and see if the other laser spot goes away. Is the second spot intense enough for you to see beads fluoresing in the stream below the nozzle? You can check the beam coming out of the front of the laser by expanding the beam with a lens onto a bit of paper, how does that look? You could then follow the beam though all the optics to the flow chamber to find out where the second beam comes from but its probably just a reflection and nothing to worry about. Simon Simon Monard Trudeau Institute 154 Algonqin Ave Saranac Lake New York 12983 Ph 518 891 3080 >>>Ray Hester <rhester@jaguar1.usouthal.edu> 04/28/05 9:40 am >>> Hi, We have a Spectra Physics 100 mW (SP model 277-GA01) argon laser on our Vantage - installed in August of this past year. Sometimes when we turn the laser on, a double spot appears on the sample stream. Other times, e.g., the next day, there is only a very nice, tight single spot. When the double spot appears, we check to make sure there are no air bubbles in the sheath fluid tubing. Also, we take off the nozzle and check to make sure there's nothing in the opening. No matter how the laser spot is focused using the beam focus knob or any of the other adjustments - X, Y, Z - the double spot won't go way. And they aren't vertically aligned. That is if you adjust the Y control knob, the two spots appear and disappear at different points - close but not one on top of the other. And this occurs when none of the other lasers are turned on. Has anyone else with this laser experienced something similar? Thanks. Ray Hester Univ. of South Alabama rhester@jaguar1.usouthal.eduReceived on Tue May 3 14:38:00 2005
This archive was generated by hypermail 2.1.8 : Sat Jan 14 2006 - 22:03:45 EST