Dear Flower, 1. Did you used single-color tubes? If not, try to prepare a single-color tube for each flurochrome you are using besides another tube for the isotypes controls. 2. On doing compensation with 7-AAD, try to use a log-scale instead of linear. Greetings Dr. M Elshal Research Fellow Flow Cytometry Core, NHLBI-NIH Bethesda, MD USA -----Original Message----- From: Kotsianidis, Ioannis [mailto:i.kotsianidis@imperial.ac.uk] Sent: Wednesday, January 26, 2005 10:29 AM To: cyto-inbox Subject: 7-AAD COMPENSATION PROBLEM Dear Flowers, I am trying to perform 7-AAD staining for cell cycle with simultaneous surface staining. When I tried to compensate 7-AAD (linear, FL3) with PE (log, FL2), 1) I wasn't able to see any event on an FL2 vs FL3 plot, while running the PE sample and 2) When I removed the FL3 signal out of FL2 while running the 7-AAD tube, I lost all the PE positivity, both in FL1 vs FL2 plot and my final sample (stained with FITC,PE and 7-AAD). I also observed the same phenomenon with APC when I tried to perform 3-color analysis with 7-AAD. Can someone please tell me what am I doing wrong? I've read the papers by Rabinovitch (1986) and also Schmid (1991) but they don't seem to worry at all about compensation problems... Thanks in advance, Dr Ioannis Kotsianidis Clinical Research Fellow Department of Haematology Faculty of Medicine - ICSTM Hammersmith Hospital Du Cane Road LONDON W12 0NN U.K.Received on Thu Jan 27 11:58:00 2005