Re: PI Staining

From: Derek Davies <derek.davies@cancer.org.uk>
Date: Tue Nov 30 2004 - 14:52:47 EST
Hi Peter,

For live dead discrimination, we add 50ul of a 50ug/ml stock to about 
a ml of cell suspension. PI max emission is around 610nm so can be 
detected in either your PE, PE-TxR or PE-Cy5 channel. Generally dead 
cells fluoresce about 2 logs brighter than live cells. On a Calibur 
you can use FL2 or FL3 - compensation is more intuitive if you look 
at it in FL2 but in some circumstances (eg very strongly expresssing 
GFP cells), FL3 may be better - you can compensate (green from red) 
via the FL3-%FL2 line.

Good luck!
Derek