Hello friends! Continuing the calibration beads theme, spherotech sent the excel template (thanks!). BUT they use the title "PMT linearity QC record"... hmmm... in order for the machine's response to be linear, shouldn't both the PMT and log-amp(s) be linear? From Prof. shapiro's book, i understand the problem is usually the log amp(s) and not the PMT(s). Using beads, is it possible to determine if it's the one or the other? Not only that, but their excel template is in 256 channels, and their doc sheet states that if i used 1024 channels, i have to convert to 256 in order to test it. Isn't that a needless "waste" of information, specially when i want to (semi)quantitate later on? Uriel.Received on Fri Oct 15 13:58:00 2004