RE: optimum concentration for sorting

From: Ann Atzberger <ann.atzberger@talk21.com>
Date: Thu Aug 26 2004 - 07:31:33 EST
Too true, also helps to watch the abort rate especially with bigger/stickier cells. I use
30psi, 100um nozzle mostly but the cell concentration depends on the cell type.
Ann Atzberger

David.C.McFarland@gsk.com wrote:

Janet, 

There is no one best concentration.  You can make some educated guesses but sometimes you
have to determine this empirically.  There are several considerations including cell
size, nozzle size, sheath pressure, sorting electronics and sample stickiness.	For
example, when doing high-speed (30,000/sec), high-pressure (45 psi) sorting with a 70
micron nozzle on a digital Vantage, the most important factor in my hands is the sample. 
For PBMCs, I request cells at a concentration of 50 to 100 million/mL and I filter
through 50 micron mesh immediately before placing on the instrument.  But this setup
would not work for an adherent cell line that clumps badly and/or that contains much
larger cells.  Contact me directly if you would like to go into more specific details
about your particular issues and I will try to help. 

Dave 

David McFarland
GlaxoSmithKline 
----- Forwarded by David C McFarland/PharmRD/GSK on 25-Aug-2004 09:14 AM ----- 
"janet dow" <jldow@unity.ncsu.edu> 
24-Aug-2004 12:09 
	To
"Cytometry Mailing List" <cytometry@flowcyt.cyto.purdue.edu> cc
Subject
optimum concentration for sorting




Dear Flowers:

What do people find as the optimum concentration of their samples for 
sorting?  I am speaking in terms of the best concentration for 
keeping the machine happy during a sort and keeping the drop delay 
constant.

Thanks in advance for all your assistance.

Janet Dow
-- 
Janet Dow
Laboratory Research Specialist and Manager
Flow Cytometry and Cell Sorting  Facility
North Carolina State College of Veterinary Medicine