From: J. Paul Robinson <jpr@flowcyt.cyto.purdue.edu>
Date: Mon Mar 22 2004 - 12:14:21 EST
I forgot to comment on the issue of file format with the FC500. 

 There is a problem with the file stucture of the FC500 files 
that we reported to Beckman-Couler a long time ago. They 
made an error in the classification of the parameter names in 
the FCS file structure. This does not have any effect if you 
analyze the data on their system, but as soon as you export 
their FCS histogram files to several other software packages 
(including Winmdi) ...its a problem becuase the parameter 
names are incorrect. (see below for a 2p histogram - both 
parmeter name data bit sizes are called $P2B - whereas the 
first one should be $P1B - so if you use another program to 
read these files, it may get confused by the double use of the 
$P2B......

further, there is no parameter in FCS 2.0 standard for word 
size with histograms - so this may cause some problems in 
defining the word size of the histogram files when you read 
them from 3rd party software programs....so if you use the 
fc500 listmode software to generate 2P histograms that are 
subsequently databased according to FCS nomenclature 
there could be a mix up od some names.

Paul Robinson
Purdue University

-----------------------header of FC500 files-------------------
FCS2.0	       128     454    1024    9216						  \$DATE\12-Mar-
04\$MODE\C\$BYTEORD\1,2\$NEXTDATA\0\$PAR\2\$TOT\
81408\$DATATYPE\I\
$P1N\SS Log\$P1S\SS Log\$P1R\64\$P2B\16\$P2N\FS 
Log\$P2S\FS Log\$P2R\64\$P2B\16\$EXP\ \
$PROJ\ \$OP\gerald\$INST\ \$CYT\Cytomics FC 500\$SRC\ 
\$CELLS\ \@LOCATION\ \
$SYS\Cytomics(TM) RXP V1.0 for 
Win32\$BTIM\08:30:08\$ETIM\08:31:13\$RUNNUMBER\4137
\ 
----------------------------------------------------------
On 2 Mar 2004 at 17:36, Stetler-Stevenson, Maryalice  wrote:

> Sorry I am late to reply on the problem with Coulter's FC500 files in FlowJo
> not showing all parameters labeled, etc. The FC500 saves two copies of the
> same data into one file.  The first data set (FCS2) is the compensated "low
> resolution" version.	The second data set (FCS3) is the un-compensated "high
> resolution" version.	Part of the problem you described is that a number of
> important keywords are missing for the "high resolution" version, making it
> difficult to use other software. We saw this problem while doing an analysis
> with WinList and notified Verity; they have added an option to address this
> issue with the FC500 files. The data looks really good with the new version.
> 
> 
> Maryalice Stetler-Stevenson
> Chief, Flow Cytometry Unit
> Laboratory of Pathology, NCI, NIH
> 
> Sometimes you're the windshield, sometimes you're the bug. 
> 


J.Paul Robinson, PhD		 PH:(765)4940757
Professor of Immunopharmacology
Professor of Biomedical Engineering  