Re: ECFP analysis on Vantage with 457nm line (secondary laserline)

From: William Telford <TelfordW@mail.nih.gov>
Date: Mon Mar 15 2004 - 15:37:18 EST
Hello Alex...

I haven't done it myself, but I'm told that the 457 nm line generated by 
the Spectrum 70C might be a little marginal for good ECFP excitation on a 
stream-in-air sorter (Coherent spec's it at 30 mW).  The labs I know of 
(including ours) that use the 457 nm line for CFP excitation generate it 
from more powerful argon-ion lasers, like the Coherent I-70C-2, -3, -4, or 
-5,  or I-90 series.  As you probably already know, CFP also excites well 
with violet lasers, either a krypton-ion source or a powerful (>20 mW) 
violet diode, if these options are available to you.

At the emission end, both of your filters will work, but you may need some 
488 nm laser blocking - 488 nm notch filters works well for this 
purpose.  If ECFP has worked poorly using your Spectrum, adding the notch 
might help in spite of the low laser power.  If you are doing CFP alone, 
you can also detect it through a GFP or FITC filter (510 - 530 nm) without 
the notch.

Enjoy,

Bill Telford
NCI-NIH

At 11:46 AM 3/15/2004 +0100, Alexander Schmitz wrote:
>Dear Flowers,
>
>On a FACSVantage SE:
>Is anybody using the 457 line of a Spectrum 70 laser  as a secondary laser 
>to excite ECFP ?
>The primary laser is 488nm.
>What is the best filter ?  470/15 , 485/22
>Do I need other filters (eg. blocking filters ?)
>Which beads can be used for the laser alignment ?
>It would be great if you could share your experiences with me.
>Many thanks
>Alex
>
>
>--
>______________________________________
>Dr. Alexander Schmitz
>FACS Core Facility
>Department for Molecular Biology (MBI) C.F. Moellers Allee 130/3
>DK 8000 Århus
>Denmark
>axs@mb.au.dk
>Tel: +45 8942 2676
>Tel: +45 2338 2429 Mobil
>Fax  +45 8916 6500
>______________________________________
>
Received on Tue Mar 16 14:38:00 2004

This archive was generated by hypermail 2.1.8 : Wed Mar 17 2004 - 03:12:03 EST