Re: PE/CY5 fluorescence problem

From: Simon Monard <smonard@trudeauinstitute.org>
Date: Thu Mar 11 2004 - 09:25:17 EST
Are you sure you are compensated correctly? If you over compensating the PE out of the
PE/Cy5 channel you would see what you describe, the brighter the PE the dimmer the
PE/Cy5. Very bright stains can also be a bit problematic to compensate especially with
log amps, you can make the CD38 stain dimmer by spiking it with unlabelled antibody. Try
running samles uncompensated then using the automatic compensation in one of the analysis
programs like FlowJo.
Best


Simon Monard
FACS Lab Manager
Trudeau Institute
154 Algonquin Avenue
Saranac Lake
NY 12983
ph 518 891 3080

Simon Monard
FACS Lab Manager
Trudeau Institute
154 Algonquin Avenue
Saranac Lake
NY 12983
ph 518 891 3080

>>> "Andrea Dewar" <andrea.dewar@imvs.sa.gov.au> - 3/10/04 3:20 AM >>>
Dear flow group,

Please see below a problem encountered by some of my colleagues. Please
reply directly to Andrew Zannettino at andrew.zannettino@imvs.sa.gov.au.

Many thanks,
Andrea


We recently performed the 3-colour flow cytometric analysis of bone marrow
(BM)derived from a patient with the plasma cell disease, myeloma, using
directly conjugated FITC and PE antibodies, and a variety of primary
antibodies
detected by sequential incubation with a biotinylated secondary and
SAV-PE/CY5
(Beckman Coulter).

The problem we have encountered centres around the fact that the level of
PE/CY5
fluorescence seen when used in isolation is greater than when it is used in
conjunction with the PE conjugated primary antibody. All our compensation
controls appear to be well set and in the range for the conjugates used.

An example of what we see is as follows: when using a mouse mAB to CXCR4 on
BM
(detected with PE/CY5), the mean fluorescence intensity (MFI) is
approximately
28.5.  However, when a PE or FITC-conjugated antibody to CD38 is used, the
MFI
of CXCR4 drops to 19.9. Is this a phenomenon related to CD38??? Or to our
PE/CY5 conjugate as the level of PE/CY5 fluorescence also dropped when
detecting other markers which we detected indirectly on these cells. Please
note that CD38 expression is extremely high on myeloma plasma cells (4th
decade
of fluorescence).

We look forward to hearing from anyone who can help.

Regards,
Andrew Zannettino
Received on Thu Mar 11 16:18:00 2004

This archive was generated by hypermail 2.1.8 : Fri Mar 12 2004 - 03:12:02 EST