Hi -

On the DiVa software, linear or log display is just that, a display 
option. All data is collected (or scaled to) 18 bit linear 
(compensated data may be broader since arithmetic operations are 
involved).

The data has the same values no matter how you display it, 2.6 to 
262K. If you display it as four decades log, the lower decade (2.6 to 
26) will be compressed into the left edge of the scale. Statistics in 
DiVa are computed on the full data range and are independent of the 
display option chosen.	Thus, even though you only display four 
decades, the mean of autofluorescent cells can be "below the bottom" 
i.e. less than 26. Choosing linear or log as a display option does 
not change this.

Hope this is clearer than mud!

Marty


>Hi,
>Would anyone who is using the BD DiVa option have worked out how to
>accurately determine a channel value on a log scale? It was easy to do in
>Analogue because it was built into the program. It is also easy to do in a
>Digital Linear scale. Any thoughts would be very helpful.
>Thank you
>Leonie
>Haematology
>POWH
>Sydney
>Australia
>
>
>Attachment converted: Asimov:InterScan_Disclaimer.txt (TEXT/ttxt) (0014C701)


-- 
Marty Bigos
Director, Flow Cytometry Core Laboratory
Gladstone Institute of Virology and Immunology

Mail:
Box 419100
San Francisco, CA 94141

Packages:
365 Vermont Street
San Francisco, CA 94103

Direct Delivery (FedEX, UPS):
Gladstone Institute
San Francisco General Hospital
Building 3 Rm 509
1001 Potrero Avenue
San Francisco, CA 94110

(tel) 415-695-3832
(fax) 415-826-8449

mbigos@gladstone.ucsf.edu
http://www.gladstone.ucsf.edu
Received on Wed Mar 10 14:58:00 2004

This archive was generated by hypermail 2.1.8 : Thu Mar 11 2004 - 03:12:02 EST