Re: granules from eo and neutrophils

From: Howard Shapiro (hms@shapirolab.com)
Date: Wed Dec 24 2003 - 17:58:55 EST


YaeJean Kim wrote:


>I have a question about granules from neutrophils or eosinophils. Does
>the flow can detect the granules released from these cells?(maybe this is
>the question regarding the lowest limit of size for particle detection by
>flow) I have purified neutrophils and eosinophils run by flow and some tubes
>activated with antigens I see little debries like group of particle with
>decreased FSC.
>
>Could they be granules? IF so, can I somehow stain the granules and see
>them?

In the 1880's, Paul Ehrlich classified blood granulocytes based on their 
staining properties. The granules in acidophils (later commonly known as 
eosinophils) were noted to have a high affinity for acid dyes such as 
eosin. Eosin is actually a fluorescein dye with halogens on the ring, and, 
as Zbigniew Darzynkiewicz is likely to mention in response to YaeJean's 
posting, fluorescein stains eosinophils very nicely provided the cells are 
permeabilized, and should therefore stain free eosinophil granules. The 
component in eosinophils that attracts acid dyes is eosinophil basic 
protein, and, if you had a labeled antibody to that, you could stain the 
granules with it; however, because the affinity of the granules for 
fluorescein is so high, you might not need the specificity provided by an 
antibody.

The granules in basophils have a high affinity for basic dyes, due to the 
granules' content of sulfated glycosaminoglycans; dyes such as the 
methylene azure dyes and toluidine blue are concentrated on or in basophil 
granules, with the result that spectral shifts (metachromasia) occur due to 
interactions between the pi orbital systems of dye molecules in close 
proximity.

Ehrlich called neutrophils neutrophils because he believed that the acid 
and basic organic dye molecules in his stain mixtures formed complexes, 
producing what he called a "neutral stain". This is not entirely correct. 
The granules of neutrophils don't have particularly high affinity for 
either acid or basic dyes, and thus do not stain strongly with either type 
of dye. If you wanted to detect neutrophil granules specifically, you'd be 
best off using an antibody to some structural component that was either 
absent from, or present in only small amounts in, both basophil and 
eosinophil granules.

The granules are probably too small to produce much of a forward scatter 
signal, although, since they are the principal contributors to the side 
scatter signal from granulocytes, you'd probably be able to pick them up by 
side scatter. For the eosinophil granules, a fluorescein signal should be 
strong enough for triggering.

-Howard