Re: 2',7'-dichlorofluorescin diacetate; Response

From: J. Paul Robinson (jpr@flowcyt.cyto.purdue.edu)
Date: Fri Nov 21 2003 - 20:19:21 EST









Well there are several issues. First, the source of the 2'7-DCFHDA is important - my own experience over many years is that the Molecular Probes reagent is the only one to use - Kodak make it and I tested a couple of lots and they were not good!

Second, yes - this reaction is well known. We published a few papers back whenever - that discuss some of the reaction issues (Measurement of intracellular fluorescence of human monocytes relative to oxidative metabolism, Robinson,J.P.; Bruner,L.H.; Bassoe,C.-F.; Hudson,J.L.; Ward,P.A.; Phan,S.H., J.Leukoc.Biol.43:304-310, 1988)

Third, there will be a continuous oxidation taking place in cell systems regardless of the stimulation you might place your cells under. Under a confocal or fluorescence microscope, you need to be careful because I believe there is some photooxidation

Finally, lymphocytes are not going to produce a whole lot of oxygen radicals - in fact, you will have a lot of trouble discriminating the background from the stimulated cells using 2'7-DCFHDA. I suggest that you use a more specific fluorochrome. In fact, I am not surprised you cannot get a stable response since the background is going to be significant.

regards
paul robinson
purdue university


-----------------------------------
On 20 Nov 2003 at 16:52, Sabyasachi Biswas wrote:

> Hi
>
> Does anyone know whether 2',7'-dichlorofluorescin diacetate (H2DCFDA)
> photodecomposes? We are trying to detect H2O2 generation in lymphocytes but our
> results are anomalous. Microscopic examination suggested that H2DCFDA decomposed
> and fluoresced when excited with FITC filter in our fluorescence microscope.
> However we couldn't find any reference for this phenomenon. Any information
> regarding this will be very helpful.
>
> Thanks
>
> Sabyasachi Biswas
>
>

J.Paul Robinson, PhD             PH:(765)4940757
Professor of Immunopharmacology
Professor of Biomedical Engineering 
Purdue University          FAX:(765)4940517
EMAIL:jpr@flowcyt.cyto.purdue.edu
WEB: http://www.cyto.purdue.edu  