Stephen Sullivan wrote:

>I am studying nuclear reprogramming in cell hybrids. I want develop a FACS 
>based system that will allow me to sort heterokaryons (fused cells where 
>the two diploid nuclei from the fused cells are still distinct) from cell 
>hybrids (where the nuclei have fused to form a single tetraploid nucleus). 
>I am currently thinking about using FRET (fluorescence resonance energy 
>transfer) to allow me so discriminate between heterokaryons and cell 
>hybrids. I plan to label two cell types with either an arbitrary nuclear 
>protein A linked to CFP or another nuclear protein B linked to YFP. The 
>nuclear two proteins will bind specifically and irreversibly to each other 
>when mixed and energy transfer should occur. So in the experiment, one 
>cell type will be labelled with nuclear protein A-CFP and the other cell 
>type with protein B-YFP. I am guessing that in heterokaryons where 
>proteins A & B will be separate and energy transfer between the CFP and 
>YFP will be minimal. In hybrids however, proteins A & B will bind and 
>energy transfer will occur. I am assuming that proteins A and B need to be 
>small and bind specifically to each other to maximise FRET. I have never 
>measured FRET before. Is this system like to work? Does anyone have any 
>suggestions how I might optimise this system. Does anyone have an 
>alternative way to sort large numbers of heterokaryons from cell hybrids ? 
>Thanks, Stephen

I wouldn't FRET about this...but, all kidding aside, I think there may be a 
somewhat simpler way to identify the hybrids. Michnick and his coworkers 
detect interactions of proteins ("A" and "B" by introducing sequences 
coding for complementary fragments of enzymes such as dihydrofolate 
reductase (DHFR) and beta-lactamase to the sequences coding for the 
proteins under study; when DHFR is used, the interaction is detected by 
binding of fluorescently labeled methotrexate. The heterokaryons should not 
bind the labeled compound, whereas the hybrids should.

Remy I, Michnick SW: Clonal selection and in vivo quantitation of protein 
interactions with protein-fragment complementation assays. Proc Natl Acad 
Sci U S A 96:5394-9, 1999

Remy I, Wilson IA, Michnick SW: Erythropoietin receptor activation by a 
ligand-induced conformation change. Science 283:990-3, 1999

Galarneau A, Primeau M, Trudeau LE, Michnick SW: Beta-lactamase protein 
fragment complementation assays as in vivo and in vitro sensors of protein 
protein interactions. Nat Biotechnol 20:619-22, 2002

You would be able to discriminate both heterokaryons and hybrids from 
single cells by DNA content using Hoechst 33342 (or possibly DRAQ5, if the 
cells will survive exposure to it), or you could use two tracking dyes, 
e.g., PKH26 and one of Molecular Probes' red-excited long side chain 
cyanine derivatives, and pick out heterokaryons and hybrids based on dual 
staining. I doubt that this will be harder than measuring energy transfer.

-Howard