I am having a problem reading APC and PerCP together when using a 4 colour
system for identifying DC subtypes in peripheral blood. Although it is
clear that I have stained successfully in both HLA-DR:PerCP and CD11c:APC, I
am having dificulty getting a filter combination that will adequately read
both APC and PerCP together.
The cytometer is: Coulter epics elite
Hi Bernadette,
It is difficult to read percp adn apc together on an Elite due to
limited compensation circuitry and cross beam excitation. You have to
measure both these fluorochromes on the same PMT using aux channels and then
can't compensate for the PE signal.
I would recommend trying to use Fitc/PE/ECD/APC as your 4 color
combination on the Elite. ECD is available from Coulter adn flouresces at
about 610nms. If you separate the laser beams in time using gated
amplification, then this combination will work.
Good luck,
Tom
****************************************************************************
*
Thomas W. Mc Closkey, Ph. D.
Director of Flow Cytometry, North Shore University Hospital
Assistant Professor of Pediatrics, New York University School of Medicine
Boas Marks Biomedical Research Center, 350 Community Drive
Manhasset, Long Island, New York 11030
ph: 516-562-4844 [office], 516-562-1135/4641 [lab] fax: 516-562-2866
****************************************************************************
*
_____________________________________________________________________
The information contained in this electronic e-mail transmission
and any attachments are intended only for the use of the individual
or entity to whom or to which it is addressed, and may contain
information that is privileged, confidential and exempt from
disclosure under applicable law. If the reader of this communication
is not the intended recipient, or the employee or agent responsible
for delivering this communication to the intended recipient, you
are hereby notified that any dissemination, distribution, copying
or disclosure of this communication and any attachment is strictly
prohibited. If you have received this transmission in error, please
notify the sender immediately by telephone and electronic mail,
and delete the original communication and any attachment from any
computer, server or other electronic recording or storage device
or medium. Receipt by anyone other than the intended recipient is
not a waiver of any attorney-client, physician-patient or other
privilege. Thank you.
This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:43:54 EST