Fw: PI staining for Apoptosis

From: Albert Tai (acktai@exelixis.com)
Date: Fri Jan 10 2003 - 17:41:55 EST


----- Original Message ----- 
From: Albert Tai 
To: cyto-inbox
Sent: Friday, January 10, 2003 12:41 PM
Subject: Re: PI staining for Apoptosis


Hi Chandra,

My experience using PI staining as an indication of apoptosis is fine but I would recommend using a secondary assay to verify your results.  The sub G1 peak you may visualized does not differentiate between apoptosis and necrosis; therefore an Annexin V type assay would help to confirm your result.  Also I will not exclude the bottom debris region as this is where you will see a sub G1 peak.  You may also try analyze your data with Modfit 3.0 from Verity software House.

Albert
  ----- Original Message ----- 
  From: chandra 
  To: Cytometry Mailing List 
  Sent: Friday, January 10, 2003 1:05 AM


  Hi
  I am doing apoptosis studies by flow cytometry using the PI stain. The PI fluroscence is being acquired in the log scale. FSC and SSC are in the linear. Right now i am unable to do the dual staining with Annexin. But i would like to know if it is fine if i produce my results only with the PI staining.
  Also i am not sure if the bottom debris region at the FSC/SSC should be included or not during the analysis.
  Thanks
  Chandr 