Re: Vantage cell counts

From: Simon Monard (smonard@trudeauinstitute.org)
Date: Mon Dec 02 2002 - 16:43:07 EST


If you sort a certain number say 30 onto a slide then count under the microscope how many
do you get? If you dont get very close to 30 then your drop delay is out of whack. If
you get close to 30 every time then some of you beads are probably sticking to the tube,
tubing etc. I wouldn't expect to see all the beads, 15000/30000 sounds about right unless
you are very careful, washing out the tube and tubing with lots of clean sheath etc
Best.

Simon Monard
FACS Lab Manager
Trudeau Institute
Saranac Lake
NY12983

Ph 518 891 3080 X352


>>> "McDonald, Ken" <ken_mcdonald@Merck.com> - 11/27/02 5:41 PM >>>

Hi All,
	I'm seeing a discrepancy in what the Vantage sorter tells me it's
sorted and the actual numbers sorted. For example, I sorted for green
calibite beads (30,000) and when I re-ran them to check for purity there
were roughly 15,000 only. Does anyone have the same problem? The purity is
good (99%) and everything else checks out by the book?..Thanks in
advance..Ken

Ken McDonald (Biologist)
Merck Frosst Canada
Tel: 514-428-2802
email: ken_mcdonald@merck.com
www.merckfrosstlab.ca


------------------------------------------------------------------------------
Notice: This e-mail message, together with any attachments, contains information
of Merck & Co., Inc. (Whitehouse Station, New Jersey, USA) that may be confidential,
proprietary copyrighted and/or legally privileged, and is intended solely for the use of
the individual or entity named on this message.  If you are not the intended recipient,
and have received this message in error, please immediately return this by e-mail and
then delete it.

==============================================================================


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:42:12 EST