RE: Fluorescence sensitivity FACScan vs. FACS Vantage SE [EV-UASEJR3]

From: David Coder (d_coder@MSN.com)
Date: Tue Nov 19 2002 - 13:44:08 EST


There are a number of differences between the two instruments, but the
most significant difference is in the fluorescence collection optics.
The  sorter uses an objective lens to image the laser/stream intercept
in air; the bench top analyzer has a similar objective lens but it is
directly coupled with a gel to the cuvette to image the laser/stream
intercept within. So, the analyzer has optics more like an oil immersion
microscope objective and thus, captures more light the free air
objective: More light means more sensitivity.

There also some lesser differences in the filters (the total number and
their transmission/reflection characteristics.) Remember that each
filter is generally no better than 90% efficient (10% of the light is
always lost), and the effect is multiplicative . That is, for three
filters (i.e.,(.9*.9*.9)) leaves you with only 75% of the light
collected by the objective lens. The FACScan has two filters in front of
the FL3 detector; the Vantage may have more depending on your
configuration. The most efficient filter for the fluorochrome may also
help. (The 650LP in the FACScan may let more Cy5 emission through the
675BP20).

Dave
===================
David M. Coder, Ph.D.
Consultant in Cytometry
Seattle, Washington
tel./message: 206-499-3446
email: d_coder@msn.com




-----Original Message-----
From: Eckstein, Volker [mailto:Volker.Eckstein@med.uni-heidelberg.de]
Sent: Monday, November 18, 2002 8:43 AM
To: cyto-inbox
Subject: Fluorescence sensitivity FACScan vs. FACS Vantage SE



Hi collegues,

is there anyone who can tell me something about the differences of
fluorescence sensitivity on the FACScan and the FACS Vantage SE?

I had a badly labelled sample of white blood cells labelled with
CD4-CyChrome (PECy5). On the Vantage equipped with a 100um nozzle it was
impossible to detect any fluorescence on FL3-channel with a 675/20
BP-filter at 100mW argon laser light. After relabelling a small aliquot
of the cells with the same antibody (10ul antibody on 4x10exp6 cells) I
could detect a dim fluorescence. Not even a log between the negatives
and the positives but good enough for a calculation. I have used
antibodies with CyChrome several times before with other cells and never
have seen such a bad fluorescence. All the optics were well aligned. But
as I said I think the amount of antibody (100ul antibody solution in a
5ml volume of PBS with 2%FCS for 2x10exp8 white blood cells) is not
sufficient for proper sorting.

Interestingly the CD4-CyChrome bad labelled cells could easily be seen
on the Fl3-channel on our old old FACScan. Even with this little amount
of antibody there was a clear and beautiful difference between CD4pos.
and neg. cells. The relabelled aliquot of cells  didn't give any
improved signal on the FACScan.

Any suggestions?

Thanks in advance

Best regards    Volker

Volker Eckstein PhD
Dept. of Internal Medicine V
Medical school of the university
University of Heidelberg
GERMANY
volker_eckstein@med.uni-heidelberg.de