Re: Internalization of antibodies

From: J.Paul Robinson (jpr@flowcyt.cyto.purdue.edu)
Date: Wed Nov 13 2002 - 09:20:37 EST


I agree with Dick Haugland - Trypan Blue has been used by Bassoe et al, to
identify internalized vs attached labeled bacteria in a phagocytosis assay. I
believe they also used Crystal Violet at one stage - although I have not tried
Crystal Violet myself.

I believe the first reference to this is
"Phagocyte C3-mediated attachment and internalization: flow cytometric
studies using a fluorescence quenching technique" Bjerknes,R.; Bassoe,C.-
F. Blut 49,315-323, 1984

Paul Robinson
---------------------------------------------------------------------------------------------
On 12 Nov 2002, at 6:23, Richard Haugland wrote:

>
> Does it have a dye on it? If it is labeled with most dyes (except
> phycobiliproteins) then trypan blue is an almost universal quencher of
> external antibodies (especially fluorescein and Alexa Fluor 488) but not
> internalized fluorescent antibodies. Useful trick for answering
> "inside-outside" questions.
>
>
>
>
>
> "Barren, Phil" wrote:
>
> > Could some one point me in the right direction?
> >
> > I have an investigator that wants to determine if their antibody is
> > internalized post binding to the surface of cells.
> >
> > Could someone offer some advice on detection, controls etc??
> >
> > Thank you
> >
> > Philip Barren