CFP "likes" a violet or blue excitation source - krypton-ion
407 or 413 nm violet, or argon-ion 457 nm blue lines all provide
not-completely-optimal but decent excitation. Unfortunately these
lines can't be had from an Enterprise II. 488 nm won't
work.
The cheapest solution for CFP on the Vantage is to obtain a violet laser
diode from Power Technologies or Coherent and set it up in the middle
(third) laser position, where your UV beam is now. These relatively
low-power (10-20 mW) laser sources provide surprisingly good excitation
of CFP. You will need to set up some sort of mounting system -
we've used a laser post mounting system from Newport. You can set
up the whole system for a good bit less than $10,000 - much less than a
water-cooled gas laser.
We have some sample data for violet laser diode excitation of CFP on our
website at
http://home.ncifcrf.gov/ccr/flowcore/index.htm
(direct access at http://home.ncifcrf.gov/ccr/flowcore/VDL.htm). If you are interested
it trying this, we can send you more specific info.
Good luck,
Bill Telford
NCI-NIH
At 12:11 PM 10/23/2002 +0100, Andrew Herman wrote:
Dear Flow-ers,
I have a colleague who wishes to analyze and sort cells that bear GFP
and CFP (cyan fluor. protein) reporter constructs.
Has anyone had experience doing this type of analysis with a Vantage SE
and an Enterprise II laser (6 color config). I realize that using a 488
nm line is suboptimal for CFP, but is it feasible?
Any info would be very much appreciated.
Thanks,
----------------------
Andrew Herman
Department of Pathology & Microbiology
University of Bristol
University Walk
Bristol BS8 1TD
Tel. +44 117 928 7511
Fax. +44 117 928 7896
A.Herman@bristol.ac.uk