RE: loader

From: Reinders, P. (HEM) (P.Reinders@lumc.nl)
Date: Tue Sep 17 2002 - 04:56:23 EST


Dear Anne,

We have intermittently using the loader for about 4 years. Several problems
did occur. The most frequent one was the jamming of the lifter rod. The
lifter rod is moved upward by a spindle. The spindle in its turn has two
ballbearings. the lower bearing can under circumstances be exposed to the
sealine fluid spilled by sample injection needle. if the driptube round this
needle is removed fluid can leak directly on the lifter rod this passes
along the sealing ring into the spindle compartment (anybody still there?).
And as most of us know ballbearings dislike sealine fluids and get jammed
eventually. We constructed a solution to divert the sealine fluid away from
the bearing. Perhaps an other fix can be found in replacing the sealing ring
by a real sealing ring: a simmer ring. We were reluctant to try this for
this might give to much friction in the rod movement. I hope this is any
help to you.


Greetings

Paul Reinders
research engineer
CKHL/Flowcytometrie E1-Q
Leids Universitair Medisch Centrum
2300 RC Leiden

071-5261672  /  071-5261636

> ----------
> From:		Husebekk Anne
> Sent:		dinsdag, 10 september 2002 10:12
> To:	Cytometry Mailing List
> Subject:	loader
>
>
> Ikke sensitiv
>
> We have had relatively well-functioning automatic loading units on two
> FacsCaliburs for some time. Since January this year they have been almost
> constantly out of order telling us that there is 'vertical motion
> time-out'.
> BD has tried to fix them several times without success.We are running
> around
> 100 samples on each loader a day.
>
> Is there anyone out there with well-functioning loaders for FacsCalibur?
>
> Anne Husebekk
> Dept. of Immunology and Transfusion Medicine
> University Hospital of Northern Norway
> 9038 Tromsų, Norway
> Phone:+47 776 26299  Fax:+47 776 26304
> Mail: anne.husebekk@unn.no
>
>