Re: CFSE labeling for in-vivo vs. in-vitro experiments?

From: Andrew Wells (adwells@mail.MED.UPENN.EDU)
Date: Fri Sep 13 2002 - 13:32:54 EST


Paul,

We've done a lot of these MLRs in vivo and in vitro using CFSE (e.g. J.I.
166:973, 2001).  You definitely use the same labeling procedure whether
you're  going to put the cells in a mouse or in a dish.  48 hours is too
early for 4/5 divisions in the in vitro (or in vivo) MLCs.  You're most
likely picking up stimulators, especially if they are autofluorescent due to
irradiation or mitomycin C treatment.  Do you see this population in your
autologous controls, too?

How are you identifying the donor/responder cells in these experiments?
Identifying them by CFSE is not enough - you need some other marker (i.e.
Thy1.1 vs. Thy1.2, Ly5.1 vs. Ly5.2, or donor-specific MHC I) to distinguish
your donor/responder T cells from your recipient/stimulator cells, both in
vivo and in vitro.  Thy1.1 B6 mice are available from Jackson (normal B6's
are Thy1.2).  Ly5 congenics are also commercially available.  Using this
type of congenic system will solve a lot of your problems.

Also, the in vitro MLRs give much more variation from experiment to
experiment (talk to anyone who does them using H3-TdR, they'll say the same
thing), and they're trickier to get working well.  The in vivo MLRs are
extremely consistent.

Hope this helps,

Andrew

--

Andrew D. Wells, Ph.D.
Assistant Professor
Department of Pathology and Laboratory Medicine
University of Pennsylvania
Joseph Stokes, Jr. Research Institute
Biesecker Liver Disease Center
The Children's Hospital of Philadelphia

phone: (215) 590-8710
fax:   (215) 590-7384

mailing address:

802 Abramson Research Center
3516 Civic Center Boulevard
Philadelphia, PA  19104


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:42:03 EST