Re: Gating for apoptosis studies

From: Derek Davies (daviesd2@cancer.org.uk)
Date: Mon Sep 09 2002 - 01:47:19 EST


Hi Dan,

You raise a good point here. When looking for Sub-G1 events you apply the
same sort of gating strategy that you would use when doing regular DNA
cell cycle analysis ie using area v width to define the singlet
population. The way I normally do it is to use the first type of gate you
show ie include all events that fall below the G1 peak and head towards
the origin - I tend to exclude the events with very low area and width - a
rule of thumb is to exclude all events below 10% of the G1 peak - this is
a fairly arbitrary cut-off but if it is applied consistently will at least
allow you to compare between samples. One thing that you have to be
careful about is calling the events in the sub-G1 region 'apoptotic'
events - all we can say is that these events have less DNA than a normal
cell, we don't know why they do - they may have undergone specific DNA
degradation via an apoptotic pathway ot they may have arrived there
through some other pathway of oncosis/necrosis. As others have said,
apoptosis is best studied not only with cytometric methods that look at
different parts of the apoptotic pathway but at different time points
following apoptotic induction. In our laboratory, we routinely look at
membrane changes with Annexin, mitochondrial membrane potential changes
with CMXRos, caspase activity and changes in cellular DNA. More
information about how we look for apoptosis on the flow cytometer may be
found by following the apoptosis link on my website.

Incidently, the smear that you see is also often found in cell cultures
that are contaminated either bacterially or by mycoplasma. A quick look
down a fluorescence microscope can be instructive!

Derek


Dan Rosson wrote:
> I'd like to ask my more learned flowlosopher colleagues a question on
> gating for the purpose of quantitating sub G1 content for apoptosis
> studies. I'm hoping Powerpoint illustrations can be attached to these
> mass mailings. We all know that we should gate out doublets and
> triplets in order to restrict a cell cycle analysis to single cells and
> one of the ways of doing this is to plot Area vs Width in order to draw
> a region. The Powerpoint attachment is a dot plot of such an experiment.
> Notice the long trail of sub G1 events that are big, fat and wide.
> Should these events be gated out as in the first histogram or left in as
> in the second histogram? It seems to me that if they are two or more sub
> G1 events are stuck together they're still apoptotic and should be left
> in the gate otherwise it's cheating the apoptotic population of its due
> representation. However, I've never seen a discussion of this.
> So, if some of you have any insight into this, I'd like to hear it.

************************************************************************
Derek Davies                       Voice: (44) 020 7269 3394
FACS Laboratory,                   FAX: (44) 020 7269 3100
Cancer Research UK,                e_mail: derek.davies@cancer.org.uk
London, UK                         mobile: 07790 604112

Web Page: http://science.cancerresearchuk.org/axp/facs/davies/index.html

In tenebris lux
*************************************************************************


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:42:02 EST