RE: Laser Power

From: Mike Evans (mevans@xyinc.com)
Date: Mon Aug 19 2002 - 18:02:13 EST


Andrew,

Yes, we see the same thing...  The laser shows up in visible and then we
change to MLUV.  We have found as much as 100mW at low power and 3 watts at
high power variance.

The photocells are calibrated based on the wavelength that you recieved the
laser in.  If you want to run 647nm then you must put a external meter and
recalibrate the photocells.

This applies to the Innova 70-90-300 systems.  On the Spectra Physics
beamlock, it is done via a 3 digit switch which changes the gain of the
photocell.

In checking the 30+ Coherent Innova 90-6 lasers that we have on sperm
sorters in the world, I believe none of them were putting out the correct
power output when they were running in MLUV.

All required photocell calibration...

Set the system power to 5 amps less than full current and then do the
photocell calibration.  This allows the laser to be linearly responsive
across the min to full current range.

If anyone needs any information about Photocell calibration, I would be
happy to offer my assistance.  I am sure Andrew would also...

regards,

Mike Evans