Re: FDA

From: Mark Kukuruga (kukuru@med.umich.edu)
Date: Tue Nov 27 2001 - 09:55:10 EST


Philippe,
First . . . as a marker for viability, calcein AM (Molecular Probes C-1430) is a better
dye than FDA, due to its better retention in cells and relative insensitivity to pH.
You can read about this (and other viability probes) at the Molecular Probes site . . .
://www.probes.com/handbook/sections/1502.html>
Second . . . related point . . . it's always good to measure membrane intactness
simultaneous with a viability probe.  Adding PI or 7-AAD will improve your live/dead
discrimination, and may help to detect artifacts due to inadequate staining or dye
efflux.
Finally . . . to answer your question . . . thymocytes are considerably smaller, and
therefore may carry far less dye.  Most likely, the signal from your 293 cells is too
bright at the concentration you're using to be effectively managed compensation-wise.
Look at lower concentrations of your viability dye.
MAK.

--
Mark A. KuKuruga, Managing Director
University of Michigan Flow Core
7416 CCGC 0946
(734) 647-3216, fax (734) 936-7376
kukuru@umich.edu


>>> Philippe Pognonec <Philippe.Pognonec@unice.fr> 11/26/01 03:58AM >>>

Hi there.
We are using FDA as a marker of living/dead cells in thymocytes preparations on a
FACSort, in FL1.
We do not observe any uncontrolable leakage of the flurescence in FL2, and are pretty
happy with it.
Fine.

No we try the same stuff on 293 cells (after dissociation following light trypsination),
and we get teh same signal in FL1, 2 and 3!!!! Impossible to use any other marker at
the same time. Does anybody have any idea what's wrong?

Philippe


____________________________________________

Philippe Pognonec, Ph.D.
Transcriptional Regulation and Differentiation
Centre de Biochimie
Parc Valrose
Universite de Nice
06108 Nice cedex 2
France
Tel/fax: (33) 492 07 64 13
http://www.unice.fr/biochimie/u470/prog5us.htm
____________________________________________


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:40:30 EST