RES: 3 color combination?

From: Esper G. Kallás (kallas.dmed@epm.br)
Date: Thu Nov 22 2001 - 11:28:46 EST


We have been using such approach without a problem, in a machine similar to
the one you have.
CD4 APC stained cells are nicely separated concurrently with PI and Annexin
V staining in FL2 and FL1, respectively.

Esper Kallas

-----Mensagem original-----
De: Nealley, Eric W Mr USAMRICD [mailto:Eric.Nealley@apg.amedd.army.mil]
Enviada em: quarta-feira, 21 de novembro de 2001 12:55
Para: Cytometry Mailing List
Assunto: 3 color combination?


We are trying to run a 3-color experiment on our FACScalibur (with FL4
option) and I was concerned about spectral overlap between propidium iodide
and APC.  We want to use annexin V FITC, propidium iodide and CD3 APC.  I
know there is a time delay procedure to calibrate the instrument, but does
the spectral overlap between propidium iodide and APC get "cancelled out" by
the time delay?  Our instrument setup:   FL1: 530/30, FL2: 585/42, FL3:
670/LP and FL4: 661/16.

Thanks in advance.

Eric Nealley