Re: Flow Woes...

From: Mark Kukuruga (kukuru@med.umich.edu)
Date: Thu Nov 15 2001 - 03:58:33 EST


Tony,
Simplest answer . . . sounds like air in the flow cell . . . have you done a drain
and fill?
If that's not it, then look for the sipper to be plugged.  With the dial in the
run position, rotate the sample tube rest into the on position WITHOUT a tube . .
you should see a drip.  Alternatively, you could remove the outer tube of the sipper
(vacuum sheath), and see the drip-back in any position.  If you don't see any drip,
try placing a piece of tubing - - one end fitting the end of the inner sipper tube,
the other connected to a syringe.  Draw back on the syringe (negative pressure)
to clear blockage in the sipper.  Be careful not to bend or torque this inner tube
. . . it goes all the way into the flow cell.
Finally, there could be crud in the flow cell, either blocking the laser, or causing
turbulence.  After a month of sitting, this could very well be the problem . . . have
you done a system bleaching recently?
MAK.
---
Mark A. KuKuruga, Managing Director
BRCF Flow Cytometry Core
The University of Michigan Medical Center
7416 CCGC 0946
1500 East Medical Center Drive
Ann Arbor, MI 48109
734-647-3216 (voice)
734-936-7376 (fax)
kukuru@umich.edu



>>> "Tony Schountz" <tschount@mesastate.edu> 11/14/01 18:39 PM >>>

I've run into a problem with our FACScan (HP system) that seems rather
bizarre. I started to set compensation manually using the FACScan Research
Software, but when I put cells on there appeared to be no flow through the
instrument (on the computer screen or the instrument panel). So I checked
all the clear lines for occlusions and none were obvious. I marked the fluid
level on the sheath tank and after a few minutes the level had dropped below
the mark, indicating that fluid was going throgh the system.

So I quit FRS and launched Autocomp and after pressing the "PMT" button I
had events occurring in the lower left quadrant of the FL1/FL2 window and
events on the FSC axis - with only water in my tube (i.e. no beads). The
flow rate of these "events" was between 3500 and 5000. Of course, after a
couple of minutes the software informs me of an error. Could someone shed a
little light on what might be occurring here? What are these "events" that
Autocomp is detecting? The instrument was used about a month ago without any
troubles, and I take great care in trying to keep the lines free of
occlusions since it frequently sets idle for this amount of time.

This instrument was donated to our college a couple of years ago, and I
don't have any formal training for maintaining it. I'd sure like to take
such a course, but it is fiscally prohibitive for me. So, I'm relying on the
collective expertise (and generosity) of the group to help me out.

Thanks,

Tony
--
Tony Schountz, Ph.D.
Dept. of Biology
Mesa State College


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:40:29 EST