Re: Miltenyi magnetic beads and CD34+

From: Paul Weiss (pweiss1@uic.edu)
Date: Thu Oct 25 2001 - 15:52:18 EST


Thank you everyone for pointing out my error.  Our lab does use 8G12
(HPCA-2) to check the purity of our Miltenyi human 34 selections.
--
Paul Weiss
Flow Cytometry Specialist
University of Illinois at Chicago
900 S Ashland Ave
Chicago, IL 60607
Phone:  312-355-0123
Fax:  312-413-7963
pweiss1@uic.edu

> From: "D. Robert Sutherland" <rob.sutherland@utoronto.ca>
> Date: Tue, 23 Oct 2001 00:46:52 +0100
> To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
> Subject: Re: Miltenyi magnetic beads and CD34+
>
>
> Hi Denice Hong and Paul Weiss,
>
> Before specifically addressing the original questions, it is important
> to note that the CD34 antibody used in the Miltenyi device is the class
> II epitope-specific monoclonal antibody QBEnd10 and NOT (as Paul
> suggested) the class III epitope-specific antibody 8G12 (commercially
> known as 'HPCA2' from BD Biosystems).  The QBEnd10 epitope is located at
> the far N-terminus of the 'flag-pole-like' CD34 molecule.  This epitope
> is composed of a linear stretch of 6 to 8 amino acid residues and as
> such is expressed on all glycoforms of the mucin-like CD34
> glycoprotein.   Again, in contrast to Paul's suggestion, it is NOT
> possible to assess the purity of the selected CD34+ cells with 'HPCA-1'.
> There are several reasons for this:
>
> First, HPCA-1 (aka clone My10), detects a class I (sialic
> acid-dependent) epitope (also in the far N-terminus) on CD34.  As such,
> it probably cannot detect all glycoforms of the mucin-like CD34 molecule
> on all cells that express them. Furthermore, since My10's binding to
> CD34 depends at least in part on the presence of specific negatively
> charged sialic acid residues, its conjugation with similarly negatively
> charged fluorochromes such as FITC efffectively destroys its already
> weak binding affinity for its epitope. Additionally, the class II
> epitope is 'nested' within the class 1 sialic acid dependent epitopes at
> the far N-terminus of the CD34 molecule and even if it were possible to
> conjugate the class 1 antibodies such as My10 with large (non-charged)
> fluorochromes such as phycoerythrin, the presence of QBEnd10 (and
> paramagnetic beads) sterically hinder the binding of large
> macromolecular complexes such as antibody:PE conjugates.
>
> So, what can be used?  Class III epitope-specific antibodies are used
> since they detect a 'common' non-glycosylation-dependent epitope of
> CD34.  The class III epitope(s) are sufficiently spatially separated
> from the class I and class II epitopes that 'orbital steering' (the
> Lunar equivalent of steric hinderance) is not a problem.   Furthermore,
> they can be conjugated with a variety of fluorochromes without loss of
> high avidity binding. Recommended reagents include clones 581 (from
> Beckman-Coulter) and HPCA-2.  These issues have been detailed is several
> publications (The CD34 antigen: Structure, biology and potential
> clinical applications. J. Hematotherapy, 1: 115-129, 1992,  The ISHAGE
> Guidelines For CD34+ Cell Determination By Flow Cytometry. J.
> Hematotherapy 3:  213-226, 1996, Structural and functional features of
> the CD34 antigen: an update. Journal of Biological Regulators and
> Homeostatic Agents 15: 1-13, 2001).
>
> We have performed several flow analyses on CD34+ cells purified by the
> Miltenyi device and as Paul indicated, the presence of the paramagnetic
> microbeads on the selected cells does not affect the light scatter
> properties (at least) of the cells.  Indeed, the Miltenyi beads are so
> small that they can only be visualised (I am told) by electron
> microscopy.  In our experience, we regularly see in excess of 97% purity
> of CD34+ cells with this device.  Since we are using this device in
> haplo-identical allo-transplant studies, we also interested in
> enumerating the residual CD3+ T cells in the selected samples.  To do
> this we perform simultaneous absolute CD34+ cell and CD3+ cell
> enumeration in a single tube using a variation of the single platform
> ISHAGE/CPC protocol (Enumeration of CD34+ Hematopoietic Stem and
> Progenitor Cells. Current Protocols in Cytometry Unit 6.4.1-6.4.22,
> 199).  Our experience indicates that very few CD3+ cells are found in
> the CD34+ fraction;  indeed data thus far shows that considerably less
> than 10e5 CD3+ cells/Kg patient weight end up in the graft!
>
> Examples (using either Beckman-Coulter or BD Biosystems instruments) of
> how these analyses are performed using Beckman-Coulter's Stem-Kit
> (CD45FITC, CD34PE and Flow Count beads) supplemented with CD3PE:Cy5  can
> be found at the ISHAGE web-site
> (www.ishage.org/committees/Committees/Graft_Evaluation/graft.htm).
>
> A few final words of caution;  if you hope to obtain any reliable flow
> data whatsoever on CD34+ cells selected with the Miltenyi device, it is
> very important to keep the selected CD34+ cell fraction cold (on ice)
> from the moment it is eluted from the column, through the staining
> proceedure and until the sample is anlaysed by flow. Failure to follow
> these simple guidelines will almost certainly result in the generation
> of large aggregates of CD34+ cells, consequent to the upregulation of
> adhesion molecules on the CD34+ cells.  The latter in turn is due to the
> fact that class II (and class I) CD34 antibodies stimulate a
> metabolism-dependent upregulation of adhesion molecule expression.  The
> presence of aggregates destroys the ability of any flow method, whether
> based upon 'two platform' (%CD34+ from cytometry multiplied by the
> absolute leukocyte count from a hematology analyser) or 'single
> platform' (bead-based absolute CD34+ cell counting) to accurately
> enumerate the selected cells.  Because of aggregates, viability
> assessments are also futile since for example, a single dead cell within
> an aggregate of cells will result in all cells in the aggregate being
> excluded from analysis. However, in the allotransplant setting, since
> the residual CD3+ cells do not get trapped in the CD34+ cell aggregates,
> it is possible, using single platform methods, to accurately assess the
> absolute CD3+ cell content in the graft.
>
> I hope this helps.
>
> D. Robert Sutherland,
> Department of Medical Oncology and Hematology
> Princess Margaret Hospital,
> The University Health Network,
> Toronto.
>
>
>
>
>