Re:Virus Sorting

From: Howard Shapiro (hms@shapirolab.com)
Date: Wed Oct 24 2001 - 20:34:14 EST


Andrew Beernink wrote-

>Has anybody sorted virus?  Any suggestions re: 1) safety, and )
>labelling/staining techniques?  I'm assuming you'd trigger off a
>fluorescence signal (Ab label), with a Hoe342 or other xNA "counterstain" to
>discount drops with multiple particles.
>
>This would be done on a MoFlo (488, I70 Spectrum, I302), if it's doable!

I recently had an e-mail exchange with Stefan Andreatta
<stefan.andreatta@uibk.ac.at>, at the University of Innsbruck, on this
topic; he was detecting (and possibly sorting) viruses on a MoFlo, and Matt
Ottenberg of Cytomation tells me that Stefan gave a nice presentation on
the subject at the European MoFlo Users' meeting a few weeks ago.

BUT-

My colleagues and I reported (in 1979) detecting scatter signals from
individual virions on a specially built flow cytometer with high light
collection, very low background, and a flow rate 1/100 of that in a
conventional flow cytometer.  The observation volume, with a 2 um beam
height and a 2 um core diameter, was only a few femtoliters.  In theory, we
should have been able to detect staining of virus by a few dozen antibody
molecules, but there were problems with the label and this was in the days
before monoclonals were available, so background fluorescence from the 99+
per cent of antibodies in a polyclonal antibody that did not react with
virus would likely have been intolerable.  The apparatus would almost
certainly have worked with phycoerythrin-labeled monoclonal antibodies;
indeed, a much simpler, conventional flow cytometer now in my lab can
reliably distinguish lymphocytes bearing just over 100 PE-labeled (1:1)
antibodies from background.
In fact, the slow flow instruments built at Los Alamos, at Twente
University in the Netherlands, and at CalTech, for sizing nucleic acid
fragments, should have even higher sensitivity; Rob Habbersett at Los
Alamos has detected single molecules of PE.

Commercial instruments are a different story.  The second paper on
detecting individual virions in flow (1999) used Molecular Probes dyes such
as SYBR-1, which enhance fluorescence several thousand times on binding to
nucleic acid; this made the virions barely detectable in fluorescence on
instruments such as the FACScan and FACSCalibur, which probably collect
light about six times as efficiently as a MoFlo.  Stefan Andreatta, as far
as I know, was using similar nucleic acid stains, and, with the
proportionally higher excitation power available on the MoFlo, apparently
managed to detect virions.  However, the particles in this case are
undoubtedly bearing at least a few thousand highly fluorescent molecules,
whereas it seems unlikely that, without really large scale fluorescence
amplification, you could get a signal detectable in a MoFlo (or even a
FACScan) from no more than a hundred or so antibodies which you would
expect to be bound to a single virion.

Hoechst dyes are not a good choice for a nucleic acid stain or counterstain
in this context; they only enhance fluorescence a hundredfold or so on
binding to DNA, and the fluorescence signal from free dye in solution would
likely swamp whatever you got from virus.

The most informative paper that I know about to date on detecting virions
in flow, again using SYBR type dyes, is Brussaard CPD, Marie D, Bratbak G:
Flow cytometric detection of viruses.  J Virol Methods 2000; 85:175-182;
these authors examined a number of viruses including bacteriophages, HSV,
and HIV.

Virus sorting, probably with a slow-flow microfluidic sorter (to eliminate
aerosol generation problems which, it now seems, apply to mail sorters as
well as cell sorters) is one of the projects I envision for my Center for
Microbial Cytometry, about which a followup to my mailing earlier this year
should appear in the coming weeks.

If Andrew can tell us what kind of viruses he wants to sort and for what
characteristics (assuming it's not a trade secret), a strategy might emerge.

-Howard