RE: Sorting on DNA Content

From: Haug, Jeff (JSH@Stowers-Institute.org)
Date: Wed Oct 24 2001 - 10:29:24 EST


Dear Charles,
	Probably the best method for DNA staining and sorting viable cell is
the method that Margaret Goodell uses for the identification of SP cells.
One place to get a very detailed procedure is Margaret's Website,
http://www.bcm.tmc.edu/genetherapy/goodell/new_site/protocols.html .
	Modify the procedure in two simple ways; 1) always use verapamil to
block Hoechst efflux, and 2) lower the PMT attenuation and gain so that G1
and G2M peaks are both on the histogram.

Best Results, Jeffrey