Hi Richard,

It is my impression that whenever a real DNA-tetraploid (G1) population is
present, these cells will further proceed through their cell cycle, with
DNA-synthesis (towards 8n DNA) and G2/M (8n DNA) being apparent.  So if you
see any events above 4n, my guess is that the 4n cells represent a happy
tetraploid G1 population, whereas if there's nothing above 4n, then your
tetraploid population is most probably a blocked G2/M population (especially
if its % is increased).

I hope this helps?
Best regards,
Dirk

Prof. Dirk Van Bockstaele
Laboratory of Hematology
Head of Flow Cytometry / Molecular Diagnostics
Antwerp University Hospital
Wilrijkstraat 10
B-2650 Edegem
Belgium
phone 32 3 821 3900 fax 32 3 825 1148


> ----------
> Van:	Salinas, Richard[SMTP:rsalinas@saci.org]
> Verzonden:	donderdag 18 oktober 2001 19:26
> Aan:	Cytometry Mailing List
> Onderwerp:	Tetraploid populations
>
>
> Greetings all,
>
> My question pertains to DNA cell cycle analysis in frozen tumors.
> By definition, tetraploid peaks have DI's from 1.9 - 2.1 with the
> proportions of cells greater than that of the normal diploid G2/M
> population.
> In samples, when the diploid S phase is usually low, I would expect a
> somewhat low G2/M as well, but when I see a G2/M of 40% I immediately
> think
> tetraploid.
> Is there a clear cut way to distinguish and validate normal G2/M cells vs.
> Tetraploid populations?
> I ask for your enlightenment.
>
> Richard A. Salinas
>