RE: Preservatives

From: Bgreig2@AOL.com
Date: Thu Sep 27 2001 - 09:28:57 EST


Dear Kathy and Lori,

  We also do lots of work for outside clients (>2k/year) and many of the samples are
  sent over the weekend.  Most of the blood and bone marrow samples are collected
  in ACD-Solution A which works really well as long as it's not partially filled.
  The pH will be too low in a short filled specimen.  Heparin is choice #2, and EDTA is
  last because of the beating it puts on myeloid elements (lymphomas don't suffer as
  much though).  Tissues, FNA's, and fluids are best preserved if kept intact and put
  in RPMI.  As long as they are not terribly necrotic, we routinely have viabilities
  >80% even after 2 or more days.  Refrigerate everything until processing.  If shipped,
  use an ice pack that is not directly in contact with the specimen.

  Even with these requirements we still work on Saturday mornings but specimen rejection
  from low viability is less than 2% based on last years' records.

  Good luck,

Bruce Greig
Flow Cytometry
Vanderbilt Univ. Med.Center